Hi has anyone removed a Picker CD 38 collimator, type R20MC? I have a unit in a vet clinic and the collimator "catches" as it is rotated. The manual does not include the collimator.

Re: CD 38 collimator [Re: Phil Coulthard] #73376
12/06/18 10:09 AM
12/06/18 10:09 AM
Joined: Feb 2004
Posts: 13,459
the path less trodden
Geoff Hannis Online content
Super Hero
Geoff Hannis  Online Content
Super Hero
Joined: Feb 2004
Posts: 13,459
the path less trodden

With a machine that old, Phil, you should count yourself lucky if that's the least of your worries!

But the problem you mention should be easy enough to identify by inspection ... slight damage, or simply worn out!

You may be able to improve matters (mechanically), but without seeing the thing it's difficult to be sure.

But ... how good are the radiographs? And does the collimator ... er, collimate (accurately)? think

Re: CD 38 collimator [Re: Phil Coulthard] #73378
12/06/18 10:14 AM
12/06/18 10:14 AM
Joined: Feb 2005
Posts: 50
Hexham
P
Phil Coulthard Online content OP
Scholar
Phil Coulthard  Online Content OP
Scholar
P
Joined: Feb 2005
Posts: 50
Hexham
The output is fine as is the field size and alignment.

The problem is dropping the collimator slowly to allow inspection. I suspect there are leaded fingers inside which are some how catching during rotation, but I have no drawings to refer to.

Re: CD 38 collimator [Re: Phil Coulthard] #73380
12/06/18 10:34 AM
12/06/18 10:34 AM
Joined: Feb 2004
Posts: 13,459
the path less trodden
Geoff Hannis Online content
Super Hero
Geoff Hannis  Online Content
Super Hero
Joined: Feb 2004
Posts: 13,459
the path less trodden

Yes; maybe some "grit" (or similar) in there.

If you're there alone (without an extra pair of hands), try resting the tube and collimator on something (pillow, chair ... whatever is available) before releasing the collimator.

Sounds like a screwdriver job, Phil. smile

Re: CD 38 collimator [Re: Phil Coulthard] #73387
12/06/18 11:31 AM
12/06/18 11:31 AM
Joined: Feb 2005
Posts: 50
Hexham
P
Phil Coulthard Online content OP
Scholar
Phil Coulthard  Online Content OP
Scholar
P
Joined: Feb 2005
Posts: 50
Hexham
No problem with the mechanics, just hoped someone had actually disassembled the device or had collimator drawing of what is inside.

Re: CD 38 collimator [Re: Phil Coulthard] #73388
12/06/18 11:41 AM
12/06/18 11:41 AM
Joined: Feb 2004
Posts: 13,459
the path less trodden
Geoff Hannis Online content
Super Hero
Geoff Hannis  Online Content
Super Hero
Joined: Feb 2004
Posts: 13,459
the path less trodden

Yes; I had gathered that. But either way you'll need to open it up to clear (clean, most likely) the fault.

So you'll know soon enough, Phil. smile

Trouble is, later on (after re-fitting) you'll probably need to fiddle about getting the alignment right again. frown

Those old collimators - or Light-Beam Diaphragms, as we used to call them - (and probably even the current ones) were much of a muchness. I have opened up a few in my time, but (again) that was all a long time ago now. You just need to be carefull.

In fact, Phil, I wish I were there with you! smile

Re: CD 38 collimator [Re: Phil Coulthard] #73400
13/06/18 09:08 AM
13/06/18 09:08 AM
Joined: Sep 2017
Posts: 63
Kent UK
I
Ian Chell Offline
Scholar
Ian Chell  Offline
Scholar
I
Joined: Sep 2017
Posts: 63
Kent UK
Morning, It's difficult to tell or remember it but going back to basics,

- you say rotation - is this internal leaves/iris or whole body rotation on the flange?

Hope you don't mind me saying but to build on Geoff's comment, to inspect the collimator action, you can compensate the weight of the collimator to allow inspection of the leaves by putting similar weighted lead-rubber aprons on the arm but also tie a rope incase they fall off (covering my back here).

BW Ian

Re: CD 38 collimator [Re: Ian Chell] #73402
13/06/18 11:12 AM
13/06/18 11:12 AM
Joined: Feb 2004
Posts: 13,459
the path less trodden
Geoff Hannis Online content
Super Hero
Geoff Hannis  Online Content
Super Hero
Joined: Feb 2004
Posts: 13,459
the path less trodden

I suspect "whole body rotation" ... but Phil isn't on site until Friday, apparently; hopefully all shall be revealed then. How about some photos, Phil?

If things are nice and quiet there, he should be able to quickly assess the situation just by putting an ear to the collimator whilst rotating it and (or) twiddling the knobs. Also by physically rotating the tube (if need be) and peering through the bottom of the collimator whilst twiddling away.

BTW Phil, you might want to check (and maybe change) the lamp whilst you're at it. If so, you'll need to carry one with you (might save another journey). smile

[Linked Image]

Meanwhile Ian, your sound advice reminds me of the time I single-handedly changed (removed, and swapped it with a "known good" one) an x-ray tube on a major Philips unit - without ropes, nor any other (safety) aids. Not in the UK, I hasten to add. But anyway, "don't try this at home, folks"!

Of course, I was much younger (not to mention, stronger) back then. rolleyes

Phil should be OK with just the collimator, though. If removal proves to be necessary, it's just a matter of being able to "catch it" when the final screw is released! A job for two pairs of hands, you might say - but hardly any weight involved (the normal locking lever should suffice to hold the tube in position).

Re: CD 38 collimator [Re: Phil Coulthard] #73408
13/06/18 09:57 PM
13/06/18 09:57 PM
Joined: Sep 2017
Posts: 63
Kent UK
I
Ian Chell Offline
Scholar
Ian Chell  Offline
Scholar
I
Joined: Sep 2017
Posts: 63
Kent UK
Geoff, not been on for a while so hi

Old fart time - I realise you're reminiscing but I wouldn't advocate any practices without safety tie off's etc - one engineer had his jaw broken by an old spring-balanced film changer that he didn't restrain.

I know what it was like then - I have much worse skeletons in my cupboard BTW.

On topic - will keep an eye out for the reply.

Re: CD 38 collimator [Re: Phil Coulthard] #73409
13/06/18 10:26 PM
13/06/18 10:26 PM
Joined: Sep 2017
Posts: 63
Kent UK
I
Ian Chell Offline
Scholar
Ian Chell  Offline
Scholar
I
Joined: Sep 2017
Posts: 63
Kent UK
Originally Posted by Phil Coulthard
Hi has anyone removed a Picker CD 38 collimator, type R20MC? I have a unit in a vet clinic


Is it on a CAT scanner?
laugh

Page 1 of 2 1 2

Moderated by  DaveC in Oz, KM, RoJo 

Who's Online Now
7 registered members (Rik, Neil Porter, Range, MikeX, JVR, Geoff Hannis, Rob Barefoot), 294 guests, and 27 spiders.
Key: Admin, Global Mod, Mod
Forum Statistics
Forums25
Topics10,211
Posts69,546
Members9,237
Most Online1,391
Mar 26th, 2018
Powered by UBB.threads™ PHP Forum Software 7.6.1.1