Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
Joined: Sep 2001
Posts: 497
Likes: 1
Umi Offline OP
Sage
OP Offline
Sage
Joined: Sep 2001
Posts: 497
Likes: 1
Hello everyone especially anaesthetic machine experts, as the crisis continues would you be able to share tips and information on the possibilities of using your anaesthetic machine to ventilate are and support a Covid-19 patient when no ventilators available and the option of using an anaesthetic machine as a last resort?


Attached Images
rose001_fig_07-01.gif (22.44 KB, 192 downloads)
Last edited by umish; 29/03/20 9:07 AM.

UMi-007

"WORK SMART NOT HARD !"
Joined: Feb 2004
Posts: 14,465
Likes: 36
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,465
Likes: 36

In short, using the diagram you have given, it won't work; it does not show the (anaesthetic) ventilator.

And anyway, why start tying up anaesthesia machines?

The last resort is, as always, the Ambu bag.

Other self-inflating manual resuscitation bags are available.


If you don't inspect ... don't expect.
Joined: Sep 2001
Posts: 497
Likes: 1
Umi Offline OP
Sage
OP Offline
Sage
Joined: Sep 2001
Posts: 497
Likes: 1
From the front line most hospitals have cut back or virtually stopped routine surgery.... so there are many anaesthetic machines not being utilised..... now you might understand where I am coming from old chap.


UMi-007

"WORK SMART NOT HARD !"
Joined: Feb 2004
Posts: 14,465
Likes: 36
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,465
Likes: 36

To be honest, no ... I don't understand.

What about emergency surgery?

But more importantly, most (if not all) anaesthetic (or "theatre") ventilators can be run without an anaesthesia machine. Gas supplies need to be consider (that is, be available) - just as they would with an anaesthesia machine - and generally an oxygen-air mixer would be needed. However, as the type (name) implies, anaesthetic (or "theatre") ventilators are generally not suited to use in an ITU situation.

As already mentioned, an anaesthesia machine by itself (without a ventilator) cannot ventilate a patient ... unless you're prepared to have someone sitting there squeezing a rebreathing bag.


If you don't inspect ... don't expect.
Joined: Sep 2001
Posts: 497
Likes: 1
Umi Offline OP
Sage
OP Offline
Sage
Joined: Sep 2001
Posts: 497
Likes: 1
We are talking about modern day anaesthetics machines with all the bells and whistles not the good old boyles....

Last edited by umish; 31/03/20 1:15 PM.

UMi-007

"WORK SMART NOT HARD !"
Joined: Sep 2019
Posts: 3
smb Offline
Newbie
Offline
Newbie
Joined: Sep 2019
Posts: 3
Hi, see attached link for GE kit.
https://www.gehealthcare.com/-/jssmedia/3c655c83bd6b427e9824994c12be0da5.pdf?la=en-us

Basically avoid low-flows, keep the fresh gas to around 4-5 lpm to avoid too much moisture, change soda lime regularly, remove vaporisers and flush system of all agent. May need to adjust alarm limits etc.

Joined: Sep 2001
Posts: 497
Likes: 1
Umi Offline OP
Sage
OP Offline
Sage
Joined: Sep 2001
Posts: 497
Likes: 1
Morning all,

Thanks for sharing SMB, would anyone have similar information for the Drager anaesthetic machine?

Last edited by umish; 01/04/20 9:42 AM.

UMi-007

"WORK SMART NOT HARD !"
Joined: Dec 2015
Posts: 110
Likes: 22
Savant
Offline
Savant
Joined: Dec 2015
Posts: 110
Likes: 22
Check your inbox.


30 years since the Chernobyl disaster and yet we still have no super heroes or zombies.
Joined: Sep 2001
Posts: 497
Likes: 1
Umi Offline OP
Sage
OP Offline
Sage
Joined: Sep 2001
Posts: 497
Likes: 1
So using an anaesthetic machine is an option but the key is to liaise with the relevant manufacturer of the device you are intending to use.

There is also a possibility of using a single ventilator on two patients....anyone can share any experience of this option?


UMi-007

"WORK SMART NOT HARD !"
Joined: Dec 2015
Posts: 110
Likes: 22
Savant
Offline
Savant
Joined: Dec 2015
Posts: 110
Likes: 22
I would stay away from the idea of placing two patients on a ventilator. It works with test lungs but humans react very differently to test lungs. I would let the clinicians decide that move and not even suggest to them beforehand.


30 years since the Chernobyl disaster and yet we still have no super heroes or zombies.
Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
3 members (vitapablo, Rickycash, brtech), 96 guests, and 16 robots.
Key: Admin, Global Mod, Mod
Newest Members
brtech, Carys, Ivan Lai, Chris.S, NimBuS06
9,998 Registered Users
Forum Statistics
Forums25
Topics10,882
Posts72,941
Members9,998
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5