Hi all, I have a problem with some ATL HDI 5000 ultrasound machines that we bought. 5 of them were considered good and the 6th one faulty before arrival. While testing each of them on arrival, I noticed they were not booting up. When I removed all the boards as directed by the service manual, it boots up to the Philips screen. When I slot in the front boards, it boots, but when I include a back board, it doesn't boot, and vice versa. This happened with all the 5 machines. I also switched their Power Supply Module (PSM) boards. Still didn't work. Eventually, I then decided to try the PSM board from the 6th faulty machine, and surprisingly, all the boards worked as I placed them one after the other. This made me feel that all the other PSM are faulty. Unfortunately, as I placed in the last board, it booted to the Philips screen and immediately switched off and started showing the OVP TRIG light.

I switched off and on but since then, it has been showing the OVP-TRIG light.

I measured the voltages on the PSM boards for both the one considered not working and the one with the OVP TRIG. Below are the readings:

For the PSM considered not working

Without boards
P2 (in order)
-14.83
14.93
-11.87
11.99
-5.94
5.97
-5.26
5.05

P3 (in order)
0.02
-0.00
1 (no continuity)
0.01
4.98
2.49
5.12
2.48

With boards
P2 (in order)
-14.79
15.26
-11.85
12.21
-5.97
6.07
-5.26
3.29

P3 (in order)
-2.25
2.04
4.25
0.00
5.04
2.51
4.97
2.51

For the PSM with OVP TRIG

Without boards
P2 (in order)
0.02
0.01
0.00
0.01
0.01
0.02
0.00
0.00

P3 (in order)
0.00
0.00
4.14
0.00
0.01
0.00
5.02
2.51

With boards
P2 (in order)
0.16
0.10
0.02
0.00
0.01
0.00
0.00
0.00

P3 (in order)
0.00
0.00
4.16
0.00
0.01
0.03
5.06
2.49

Pls, I need help fixing this problem. Any help will be appreciated.

Kind regards,
Emeka

Philips
Re: Need help fixing ATL HDI 5000 ultrasound machine
Mekaz #75650 03/09/20 6:39 PM
Joined: Feb 2004
Posts: 14,165
Likes: 13
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,165
Likes: 13
OVP TRIG indicates that +6 V or -6 V have higher than normal voltage, or -5.2 V current limiting has been activated (-5.2 V circuits are drawing more than 14.3 A).

Sounds like you have an interesting situation there, Emeka; but at least you have enough machines to play with.

Normally swapping boards about like that is not recommended. If "needs must", swapping one at a time may be required, but "with caution". But who knows what went on before you laid hands on them!

From what you are saying (and at first glance), it sounds like all but one PSM is faulty, as well as the "last board".

I'm guessing a possible short circuit on the "last board" (as that would presumably activate the OVP TRIG current limiting indication).

I see that you have the manual ... is it any help?


If you don't inspect ... don't expect.
Re: Need help fixing ATL HDI 5000 ultrasound machine
Geoff Hannis #75651 03/09/20 7:01 PM
Joined: Sep 2020
Posts: 7
Mekaz Offline OP
Newbie
OP Offline
Newbie
Joined: Sep 2020
Posts: 7
I am thinking the OVP-TRIG may be as a result of frequency difference. The machine requires 100 - 120V; 60Hz but our electricity in Nigeria is 220 - 240V; 50Hz. I used a stepdown transformer to reduce the voltage but I don't think it also converts the frequency to 60Hz.

What do you think about it?

Re: Need help fixing ATL HDI 5000 ultrasound machine
Mekaz #75652 03/09/20 7:15 PM
Joined: Feb 2004
Posts: 14,165
Likes: 13
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,165
Likes: 13
I believe that machine should be able to handle 50/60 Hz OK ... but it won't be too happy with any power interruptions

By the way, those power voltages you have listed look OK to me.

Keep digging.


If you don't inspect ... don't expect.
Re: Need help fixing ATL HDI 5000 ultrasound machine
Geoff Hannis #75653 03/09/20 8:27 PM
Joined: Sep 2020
Posts: 7
Mekaz Offline OP
Newbie
OP Offline
Newbie
Joined: Sep 2020
Posts: 7
I really am stuck at this point. I was thinking I could get a guide on troubleshooting and repairing the PSMs.

Re: Need help fixing ATL HDI 5000 ultrasound machine
Mekaz #75654 03/09/20 8:49 PM
Joined: Feb 2004
Posts: 14,165
Likes: 13
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,165
Likes: 13
It's difficult to advise further without actually being there.

But don't lose hope ... at least you have six machines, rather than just the one - so you can make comparisons when needed.

Just proceed in a patient, logical manner. Yes; power supplies first (as everything else depends upon those voltages).

With ultrasound units such as the HDI-5000 (which is an excellent machine), faults are usually (mostly) power supply related, or (less often) hard-drive related.

At first, try to get one machine up and running. You know, one step at a time.


If you don't inspect ... don't expect.
Re: Need help fixing ATL HDI 5000 ultrasound machine
Mekaz #75655 04/09/20 3:50 PM
Joined: Feb 2004
Posts: 14,165
Likes: 13
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,165
Likes: 13
Regarding any possibility to convert to 220 V working, you shall need to take a look at the AC Input Module (ACIM).

I believe there have been various versions of ACIM to suit specific mains power conditions; so your options will depend on which ACIM's you have.

I wondering now if all six are the same!


If you don't inspect ... don't expect.
Re: Need help fixing ATL HDI 5000 ultrasound machine
Mekaz #75656 04/09/20 4:02 PM
Joined: Oct 2006
Posts: 228
Likes: 7
Master
Offline
Master
Joined: Oct 2006
Posts: 228
Likes: 7
Hi Mekaz, I can send you two service manuals and a diagnostic supplement for the HDI 3000, as they are the same family they be may be of some use.

Send me a private mail and I'll "Wetransfer"

Stay Safe

Malcolm.

1 member likes this: Huw
Re: Need help fixing ATL HDI 5000 ultrasound machine
Geoff Hannis #75667 11/09/20 10:39 PM
Joined: Sep 2020
Posts: 7
Mekaz Offline OP
Newbie
OP Offline
Newbie
Joined: Sep 2020
Posts: 7
I have ordered a new ACIM (230/240V; 50Hz). I shall use it to commence testing.

Re: Need help fixing ATL HDI 5000 ultrasound machine
Malcolm #75668 11/09/20 11:03 PM
Joined: Sep 2020
Posts: 7
Mekaz Offline OP
Newbie
OP Offline
Newbie
Joined: Sep 2020
Posts: 7
Thanks Malcolm. Mail sent.

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (Huw), 506 guests, and 29 robots.
Key: Admin, Global Mod, Mod
Newest Members
clfs12, John75, HuyNguyen, Sahibnoor, Stuart Eccles
9723 Registered Users
Forum Statistics
Forums25
Topics10,616
Posts71,816
Members9,723
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.4