This T300 analyzer,Cannot display zero position.
I have checked the sensor and the op amp circuit and found no faults.
I don't know how to adjust it.

If you know how to adjust please let me know, thanks
[Linked Image from ftp.bmp.ovh]
[Linked Image from ftp.bmp.ovh]


daisizhou#sina.com (#=@)
Recruitment / Jobs
Advertise your jobs with EBME
Joined: Feb 2004
Posts: 14,661
Likes: 62
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,661
Likes: 62
I believe you need the software (and connect to a PC) in order to re-calibrate and reset to zero.

You probably need a new oxygen sensor, as well.

Have you hounded down the software yet?


If you don't inspect ... don't expect.
Joined: Oct 2011
Posts: 192
Likes: 10
Mentor
OP Offline
Mentor
Joined: Oct 2011
Posts: 192
Likes: 10
Yes, I have not found the software,I disassembled the main unit and checked the circuit board.
Below are some pictures I took apart, hoping to give me repair suggestions
[Linked Image from ftp.bmp.ovh]
[Linked Image from ftp.bmp.ovh]
[Linked Image from ftp.bmp.ovh]
[Linked Image from ftp.bmp.ovh]
[Linked Image from ftp.bmp.ovh]
[Linked Image from ftp.bmp.ovh]


daisizhou#sina.com (#=@)
Joined: Oct 2011
Posts: 192
Likes: 10
Mentor
OP Offline
Mentor
Joined: Oct 2011
Posts: 192
Likes: 10
I checked the data sheet of the sensor,VCC is 5V power supply,vout is the analog output voltage.
For example,0 cmh2o=2.92v,10cmh2o=3.15v,20cmh2o=3.38v,30cmh2o=3.61v.........0.2xV voltage increase.The highest I chose 4.92V, the pressure reached 85.07cmh2o.

But the low pressure range can only reach 85.07cmh2o,The instructions say that it can reach +-120cmh20,I do not know why
Low pressure can reach as low as -128.05cmh2o,I think this is normal


daisizhou#sina.com (#=@)
Joined: Oct 2011
Posts: 192
Likes: 10
Mentor
OP Offline
Mentor
Joined: Oct 2011
Posts: 192
Likes: 10
[Linked Image from ftp.bmp.ovh]

I am at the pressure sensor voltage output interface,Use a separate power supply to connect the voltage.
View the pressure value on the screen and record the corresponding voltage value
But why the pressure value can only reach 85.07cmh20and 85.44psi?
According to the manual, it should be +120cmh2o and 100psid


daisizhou#sina.com (#=@)
Joined: Oct 2011
Posts: 192
Likes: 10
Mentor
OP Offline
Mentor
Joined: Oct 2011
Posts: 192
Likes: 10
After 6 months of testing, I finally finished the calibration of the T300 analyzer, thank you friends for your help
[Linked Image from i.postimg.cc]


daisizhou#sina.com (#=@)

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
7 members (Geoff Hannis, mosfet1996, Yousri, Neil Porter, carl ray, Huw, 1 invisible), 609 guests, and 14 robots.
Key: Admin, Global Mod, Mod
Newest Members
Yousri, mosfet1996, rajvenugopal, Arzo Momand, steve_shomz
10,180 Registered Users
Forum Statistics
Forums25
Topics11,063
Posts73,728
Members10,179
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5