ah, the 30 day window

I've always allowed (and implemented) a 30 day window, literally forever, I've never seen any national guidance giving me authority however. But this is still what I practice and preach, its down to local policy.
There's a few reasons, both of which are entirely justifiable and logical. My view is that your backlog measurement trigger interval should match your tolerance for PPM slip.
[BACKLOG]
I have always classed backlog as being 'Those devices more than 1 month outside of their planned date'. We have regular reviews of the backlog list to ensure we are keeping it minimal and safe.
This gives staff and the teams up to a month to physically locate the equipment and send it to Clinical Engineering before we start to worry too much.
[Location Based Fixed PPM]
Location Based PPM: When undertaking location based PPM on large departments and completing the whole area, it could take anywhere from 1-3 weeks depending on the size of the department. the tolerance of 1 month helps to ensure that the department your visiting daily with the extended team during PPM doesnt trigger unnecessarily onto the backlog list.
[Rolling PPM]
When undertaking rolling planned maintenance on equipment types such as incubators, anaesthetic machines and the like, these often are notoriously difficult to release due to shift patterns, usage and the fact that they are high in demand, difficult to transport etc. Its a reasonable measure to allow some tolerance for the clinical staff to 'Plan' or schedule the release of the equipment.
We have equipment leads for these types of equipment (rolling) who communicate with the clinical staff/housekeepers to ensure the equipment comes back to us in an appropriate manner.
.
Across the whole of the system, both rolling and location based, we have automated reports sending reminders etc.
We monitor carefully the 'Backlog' and 'High Acuity Rolling PPM' for any which are hitting the end point which forms a very reasonable response system given the fairly tight tolerance of 1 month. Were talking low numbers but with high intervention for the size of the trust we are!
If we had to shift this downwards to 3,2,1,0 weeks etc. We would be crying wolf all the time to ourselves and may end up with an 11 month schedule rather than a 12 in order to make it work without compromise.
An 11 month schedule would generate a never ending spiral of resourcing issues and re-adjustment of a steady fixed programme.
We dont have the resource to undertake an 11 month schedule, its hard enough for us all to to keep things moving on a 12 month one.
My advice is to include a definition of backlog in your medical devices policy, not just for KPI purposes but for risk appetite as this may be not only reasonably practicable, but sensibly implementable.
I've known/heard of some trusts not classifying backlog until it hits the 2,3,4 month stage, or allocating maintenance in 2,3 month windows. well if this is the case your tolerances/appetites for risk should be adjusted accordingly.
Its an interesting subject, everyones got their take in some way, in terms of practical application on a large scale, I've made my bed and have to live by it.