This is an analyzer of my friend. He bought it 5 years ago and it has been working fine. The firmware version is 1.01.
Now this host has a problem. The problem is that the pressure cannot be returned to zero. In simple terms, it cannot be used.

I disassembled the host. Since it is beyond the manufacturer's warranty period, if I want to repair it now (I mean looking for FLUKE service), I need to pay additional service fees.

I disassembled it in the hope that friends who buy the same type of host will be cautious in purchasing, because the quality is not reliable.
In addition, I have not been able to find the entrance to VT650 calibration. If you have the same interest, please contact me. Thank you.[Linked Image][Linked Image]
[Linked Image]


daisizhou#sina.com (#=@)
Joined: Oct 2011
Posts: 237
Likes: 20
Master
OP Offline
Master
Joined: Oct 2011
Posts: 237
Likes: 20
[Linked Image from i.postimg.cc]
[Linked Image from i.postimg.cc]
[Linked Image from i.postimg.cc]


daisizhou#sina.com (#=@)
Joined: Oct 2011
Posts: 237
Likes: 20
Master
OP Offline
Master
Joined: Oct 2011
Posts: 237
Likes: 20
[Linked Image from i.postimg.cc]
[Linked Image from i.postimg.cc]
[Linked Image from i.postimg.cc]


daisizhou#sina.com (#=@)
Joined: Oct 2011
Posts: 237
Likes: 20
Master
OP Offline
Master
Joined: Oct 2011
Posts: 237
Likes: 20
VT650 frequently experiences respiratory ratio display errors(I:E).
Sometimes it may not trigger,Need to manually adjust the threshold,Only then will it turn green normally and trigger.

This is a problem encountered in practical use,FlukeBiomedical does not have a complete solution, and the official response is that it will be resolved through firmware upgrade, but it has not been perfected yet


daisizhou#sina.com (#=@)
Joined: Jul 2007
Posts: 29
Dreamer
Offline
Dreamer
Joined: Jul 2007
Posts: 29
Hi ! did you:
- replaced the O² sensor cell (seems to be from 2019)?
- try to zero the pressure sensors WITHOUT the caps on the inlets ?

Best of luck

Joined: Oct 2011
Posts: 237
Likes: 20
Master
OP Offline
Master
Joined: Oct 2011
Posts: 237
Likes: 20
Thanks.

I tried replacing the O² sensor unit and pressing the zero button,But unfortunately, there is no improvement.
The most important issue is that users cannot perform calibration operations privately.
I don't even know how to get into calibration(service) mode

I try to read the FM24CL64BG chip,It seems that the calibration information is not stored internally,I guess the calibration information should be stored inside the main chip LPC1857FET256

Attached Images
FM24CL64BG.zip (227 Bytes, 8 downloads)
SHA1: de11a0963339a9315763bed1696751785e5df1e5
Last edited by daisizhou; 30/12/24 6:35 AM.

daisizhou#sina.com (#=@)

Moderated by  Huw 

Link Copied to Clipboard
Who's Online Now
2 members (mini nair, mub777), 643 guests, and 17 robots.
Key: Admin, Global Mod, Mod
Newest Members
Richard 1x1, Stephen UHM, AndyC85, Jose Luis, jaringa
10,258 Registered Users
Forum Statistics
Forums25
Topics11,164
Posts74,190
Members10,258
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5