Hi

Does anyone use the above in their trust?

If so, I have a question.+

Joined: Dec 2015
Posts: 163
Likes: 30
Mentor
Offline
Mentor
Joined: Dec 2015
Posts: 163
Likes: 30
There's a manual to download at the bottom of the page. https://inspirationhealthcaregroup.com/product/lifestart%EF%B8%8F/


30 years since the Chernobyl disaster and yet we still have no super heroes or zombies.
Joined: Oct 2006
Posts: 401
Likes: 11
Neoteny Offline OP
Sage
OP Offline
Sage
Joined: Oct 2006
Posts: 401
Likes: 11
I already have a manual, thanks.

Joined: Oct 2022
Posts: 63
Likes: 9
Scholar
Offline
Scholar
Joined: Oct 2022
Posts: 63
Likes: 9
Hi buddy, yes we have a couple of these here

Joined: Oct 2006
Posts: 401
Likes: 11
Neoteny Offline OP
Sage
OP Offline
Sage
Joined: Oct 2006
Posts: 401
Likes: 11
What size cylinders do you use?-

Joined: Oct 2022
Posts: 63
Likes: 9
Scholar
Offline
Scholar
Joined: Oct 2022
Posts: 63
Likes: 9
ahhhhh...now THERES a story. They purchased these around 4 years ago as they simply had to have them (2 in total) with no consultation with the med eng dept. When they came to use them, the cylinders specified were not listed via BOC so they were given the option to purchase cylinders from an alternative supplier which they did not do. A couple of months ago they asked me If i could modify them so an E size PI cylinder would fit. The cylinders are way too large to fit safely plus they had used the MODIFY word so this wasnt possible. They had purchased a 4 year service contract with them so inspiration were coming in every 12 months to unwrap them, service them then wrap them up again !! When i queried this as the contract renewal was due, the ward staff advised me that they could now do cord connected care using the atom units so they no longer needed them and that maybe another dept in the trust might be able to use them! Im not sure where else they may be used other than the delivery suite or NNU so as such they are now in storage having never been used. I take it you are having issues with cylinders also?

Joined: Oct 2006
Posts: 401
Likes: 11
Neoteny Offline OP
Sage
OP Offline
Sage
Joined: Oct 2006
Posts: 401
Likes: 11
Hi,yes, just as you have said. We were asked the question, the cost of getting Air is rediculous.

Joined: Nov 2005
Posts: 160
Likes: 4
Mentor
Offline
Mentor
Joined: Nov 2005
Posts: 160
Likes: 4
Hi

From memory when I worked there, it was a size AZ air cylinder which fitted, it was possible to get via BOC but I have no idea on the costs involved, sorry

Mark


Mark Radbourne
I work for Löwenstein Medical in the UK
1 member likes this: carl ray

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (Huw, daisizhou), 11,857 guests, and 146 robots.
Key: Admin, Global Mod, Mod
Newest Members
AraEmin, Edouardohaag, Mark B, Royal, davidarhui
10,392 Registered Users
Forum Statistics
Forums26
Topics11,281
Posts74,584
Members10,392
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5