Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 3 of 3 1 2 3
Joined: Oct 2011
Posts: 250
Likes: 22
Master
Online Content
Master
Joined: Oct 2011
Posts: 250
Likes: 22
Hi MikeSzym

What specific test are you doing? I'm interested in this.


daisizhou#sina.com (#=@)
Recruitment / Jobs
Advertise your jobs with EBME
Joined: Sep 2020
Posts: 3
Newbie
Offline
Newbie
Joined: Sep 2020
Posts: 3
The following section of IFU, but I wanted to confirm the Drager setup they use:

Check safety features:
– Correct functioning of the emergency expiratory valve: Pressure rise 1.9 to 4.4 mbar (1.9 to 4.4 cmH2O) at a flow of 4.5 to 5.5 L/min.
– Correct functioning of the non-return valve in the expiratory valve.
– Correct functioning of the emergency breathing valve: Maximum pressure drop of 4 mbar (4 cmH2O) at a suction flow of 60 to 65 L/min.
– Correct functioning of the non-return valves in the gas inlet for O2 and Air

Joined: Oct 2011
Posts: 250
Likes: 22
Master
Online Content
Master
Joined: Oct 2011
Posts: 250
Likes: 22
Usually, as I recall, during the power-on self-test, the test items you described can be completed.

If you are particularly concerned about the accuracy of the flow rate, you need an gasflow analyzer to accurately test

Looking forward to your test results


daisizhou#sina.com (#=@)
Joined: Sep 2020
Posts: 3
Newbie
Offline
Newbie
Joined: Sep 2020
Posts: 3
Thanks mate,

But those particular tests require a more nuanced setup ... i.e. injectors and creation of certain suction levels.

Joined: Oct 2011
Posts: 250
Likes: 22
Master
Online Content
Master
Joined: Oct 2011
Posts: 250
Likes: 22
Don't you have a dedicated gasflow analyzer? For example, Fluke VT650?VT PLUS HF?Timeter RT-200?T300.....


daisizhou#sina.com (#=@)
Page 3 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (daisizhou, 1 invisible), 2,125 guests, and 44 robots.
Key: Admin, Global Mod, Mod
Newest Members
Kurran, Carwyn Williams, Tony_ Parsons, Tahmores, LUISZ
10,291 Registered Users
Forum Statistics
Forums25
Topics11,186
Posts74,264
Members10,290
Most Online5,980
Jan 29th, 2020
Powered by UBB.threads™ PHP Forum Software 7.7.5