SUL CS-2400 - Sysmex Coagulation Analyser - 20/02/19 4:34 PM

Hi, would anyone know if it is OK to perform a standard Electrical Safety Test on this device?
I seem to remember some one telling me that testing this may cause some damage.
Thanks.
Posted By: Geoff Hannis

Re: SUL CS-2400 - Sysmex Coagulation Analyser - 20/02/19 9:02 PM


It depends what you mean by a "standard" Electrical Safety Test - the one for kettles? think

If you follow the appropriate Standard for equipment of this type, then all should be well.

I believe you need to look out for BS EN 61010-1. smile

Electrical Safety Testing to that Standard is simple (and limited) ... all you need is one of these* (see Appendix C at page 32).

* Other suitable testers may be available.
Posted By: Bri1965

Re: SUL CS-2400 - Sysmex Coagulation Analyser - 21/02/19 9:13 PM

Thanks very much Geoff.