Earth leakage current for safety test fail - 06/01/09 6:46 AM
Hye friend,
I doing a safety test for one unit of CO2 laser, modeled QrayFRX. The test standard I used was IEC601-1, and I put under class 1, type BF.
The results that I got was failed. When I refer to the results, the measurement that failed are;
Earth leakage current NC - (limit value=0.5)- (result=1.704mA)
Earth leakage current SFC - (limit value=1.0)- (result=2.627mA)
How come this test can failed, what are the condition that can contributes to it....
I really hope anybody can helping me resolve this matters.,,,
I doing a safety test for one unit of CO2 laser, modeled QrayFRX. The test standard I used was IEC601-1, and I put under class 1, type BF.
The results that I got was failed. When I refer to the results, the measurement that failed are;
Earth leakage current NC - (limit value=0.5)- (result=1.704mA)
Earth leakage current SFC - (limit value=1.0)- (result=2.627mA)
How come this test can failed, what are the condition that can contributes to it....
I really hope anybody can helping me resolve this matters.,,,