Posted By: arman andaya A.C. Line filter - 10/03/11 3:11 AM
Good day to all,

I need assistance,first what will happen to a medical unit(x-ray)
if the line filter is defective? second, How to test/check bad or good line filter?

thanks
Posted By: Geoff Hannis Re: A.C. Line filter - 10/03/11 6:58 AM

Welcome to the forum Arman. smile

Is this a hypothetical question, or do you have an actual faulty unit?

If so (and assuming we are talking about somewhere near the incoming mains connections of the unit), are you sure you are looking at a line filter ... or is it some other component, like a transformer of some sort, or line resistance?

What kind of x-ray unit are we talking about here? Portable, mobile, or a fixed unit? Single, or three phase power input?
Posted By: arman andaya Re: A.C. Line filter - 10/03/11 9:33 AM
shemadzu x-ray,its a fixed one and single phase input supply.
this is EMF/EMI filter some called line filter(or whatsoever).connected after circuit breaker and magnetic contactor.
Posted By: Geoff Hannis Re: A.C. Line filter - 10/03/11 9:53 AM

What's it like? A "box" ... or toroidal "doughnut" or some such thing?

Do you suspect that it is faulty? smile
Posted By: arman andaya Re: A.C. Line filter - 10/03/11 10:03 AM
a box type about 3"x6"x8".
and as you said ,I suspect that.
Posted By: Geoff Hannis Re: A.C. Line filter - 10/03/11 10:29 AM

Is the machine not working (not powering up)?

Why not see if you can (safely) bypass the filter box, in order to confirm (or eliminate) that as the faulty element? smile
Posted By: arman andaya Re: A.C. Line filter - 10/03/11 11:54 AM
the x-ray are o.k. sometimes,when their is sufficient power( I think)and sometimes when I expose there is no exposure but the rotor is turning.what will I do?

I already by pass the filter,and the program(anatomical) corrupted. cry
Posted By: Geoff Hannis Re: A.C. Line filter - 10/03/11 2:52 PM

You're saying that the anatomical program is corrupted.

Does the machine work OK when factors are set manually?

Perhaps it is the APR program that is inhibiting exposures. smile