Endoscopy system installed on medical pendant - 05/07/16 9:43 AM
Need all expert to advice and thanks in advance.
Do I need to have a isolation transformer if the individual equipment in my endoscopy system meets IEC60601-1 standards. In addition, they are all plugged to individual power socket on the pendant with 10mA RCCB.
I understand that for Endoscopy system installed on mobile trolley with ganged-socket an isolation transformer is required on the trolley so that the leakage current is keep within safe limit.