It's always interesting to watch the IDA graph when testing the occlusion pressure (but, as you have indicated, very difficult to "see" this from a printout of figures, or simply observing the digits on the tester itself). Pumps with peristaltic mechanisms may work fine at normal pressures (and the users may never complain), but (as I say) sometimes they won't "hold the pressure" and/or never attain enough pressure to reach the occlusion alarm.
Myself, I generally tend to look at the occlusion pressure first. If it reaches alarm levels, then at least the pump is "man enough", as it were. If that's OK, then I take a look at the average flow rate and its inseparable friend, volume delivered. Following that, of course, is when the "fun" begins. That is, getting the flow rate accurate (which is usually easy enough with many - but not all - volumetric pumps) ... and then, the occlusion pressure (easy on some pumps, "not so easy" on others).
Hey ho. As I say, (many) happy hours of fun!