Hi Folks
we have the Infusomat infusion pumps at our hospital and we are having major problems infusing blood. The problem is when infusing blood the accumalative air alarm keeps activating.i am dealing with Braun concerning this problem and their atitude is that our staff are not following the correct proceedures and there is nothing wronge with the pumps.

Are any of you having the same problem?
if yes what is being done to correct the problem?

your comments are apprieciated.

richard

Joined: Apr 2003
Posts: 91
Adept
Offline
Adept
Joined: Apr 2003
Posts: 91
Hello Richard,

One of our departments here had exactly the same problem.

Talks with Braun revealed the problem is with the giving set and not the pump, therefore Braun are in the process of producing a new giving set primarily for giving blood.

Nothing to do with your staff or procedures whatsoever.

Hope this helps

Rob

Last edited by Skippy; 06/01/10 2:03 PM.

Live and Let live.
Joined: May 2003
Posts: 75
Adept
OP Offline
Adept
Joined: May 2003
Posts: 75
Hi Rob

thanks for your reply
what hospital do you work for?

regards

richard

Joined: Aug 2007
Posts: 145
Expert
Offline
Expert
Joined: Aug 2007
Posts: 145
We have special b braun pumps for blood on the paediatric unit and different sets. Not sure how the pumps are different to the normal ones, they look the same, I think it is in the programing.

Joined: Oct 2002
Posts: 74
Scholar
Offline
Scholar
Joined: Oct 2002
Posts: 74
Hi Richard
We have a good number here in Edinburgh and have set the accumulative alarm to the max of 3.8ml/hr. Have you done the same? In saying that we still have alarms regarding this. At the moment I am looking into a Datix report on this very point but similar to everyone else I am sooooo busy!
Dave D

Joined: May 2003
Posts: 75
Adept
OP Offline
Adept
Joined: May 2003
Posts: 75
Hi Dave

What is the Datix report?
is there any chance you can send me a copy or let me know where i can get it?
we have set all our pumps to 3.8ml per hour but the problem is persistant.

regards

richard

Joined: Jan 2010
Posts: 4
Newbie
Offline
Newbie
Joined: Jan 2010
Posts: 4
Hi, We have these pumps and the Blood giving sets. The problem is also when giving TPN. I have been in contact with Braun loads of times also gave more indepth training to our staff here (teaching some of them was like teaching granny to suck eggs) but i covered all bases regarding Braun saying it was user error!. However we still do have a problem sometimes. We have cranked up the setings as much as poss but still probs. I was told a while ago that some new software might have to put on them .... but as yet i havnt heard anymore.
so now i just hand over any 'faulty' pump to the boys at meed
and smile.

Last edited by magallan; 21/01/10 10:51 AM.
Joined: Oct 2002
Posts: 74
Scholar
Offline
Scholar
Joined: Oct 2002
Posts: 74
Richard
"Datix" is our in-house incident reprorting system. I would be unable to send any reports because of that, sorry.

Dave

Joined: May 2003
Posts: 75
Adept
OP Offline
Adept
Joined: May 2003
Posts: 75
Hi All

i am meeting with B-Braun in two weeks concerning our problems with blood transfusions.If anybody else out there has the same problems with bood/tpn infusions i would like to hear from you also if you use th B-Braun and have no problems with blood/tpn infusions.the more info i get the better position i will be in when we meet B-Braun.

regards

richard


Joined: Jan 2010
Posts: 4
Newbie
Offline
Newbie
Joined: Jan 2010
Posts: 4
Hi Richard,
I have started an audit of my 'faulty' pumps, so can keep you in touch with my findings.
Good luck with Braun and any different answers they come up with I will definatly like too hear if thats ok?

Regards Liz.


Page 1 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 1,152 guests, and 23 robots.
Key: Admin, Global Mod, Mod
Newest Members
Nick.A, Tdayve, AdrianT82, Iagonuns, Dctm
10,371 Registered Users
Forum Statistics
Forums26
Topics11,261
Posts74,526
Members10,371
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5