Hi, All Friends
I have a Incubator Drager 8000 SC have a voice too high a motor during operation .
Should I change the motor, this is not possible to .
Or what do I do so there is no sound from the motor .


Thanks,

Joined: Apr 2009
Posts: 81
Adept
Offline
Adept
Joined: Apr 2009
Posts: 81
I have similar problem in my 8000c.
but this is not a motor problem.
I found just a very small fan which make a noise.

Joined: Aug 2007
Posts: 145
Expert
Offline
Expert
Joined: Aug 2007
Posts: 145
Try cleaning the part were the silicon node attaches to the fan. Often these kinds of problems are sorted out if you take it apart and put it back together again, even if you don't locate the fault.

Joined: Feb 2004
Posts: 14,807
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,807
Likes: 72

For what it's worth ... only the other day a Wise Man told me that it was "well known" that Dräger engineers disconnected the noisy on-board fan. frown

My own take is that (considering the price of the incubator) we might have expected something a little more er, ... durch Technik!


If you don't inspect ... don't expect.
Joined: Apr 2008
Posts: 101
wensh79 Offline OP
Savant
OP Offline
Savant
Joined: Apr 2008
Posts: 101
Hi,
Deenis - I removed this fan has already been changed and still sound high motor .
Webbie - Has been cleaned and the fan motor well, and the sound is still present ( not a high as it was ) .
Geoff - I do not understand exactly what you mean, please explain more


Thanks for All,

Joined: Feb 2004
Posts: 14,807
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,807
Likes: 72

You need to establish (and make clear) which fan motor you are talking about, Hassan.

The fan motor (large motor, main circulation fan) ... or the PC-processor style cooling fan on the circuit board.

If the latter ... if it were me, I would consider disconnecting it. smile

This German phrase Vorsprung durch Technik is usually translated into English as "progress through technology".

It has been used in advertising Audi cars since the 1980s, capitalizing on the German reputation for technical expertise.

Although, in the case in question, perhaps the (older) English saying about spoiling the ship for a ha'p'orth* of tar may be more appropriate.

Essentially that means that if you have generally done a good job on something, you should keep up the same high standards throughout.

* Half-penny's worth.


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,807
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,807
Likes: 72

Wise Man II informs me that the small PCB-mounted fan is needed only if RS-232 is enabled (which it was not in the UK). Hence the "drill" in the UK was to cut the wires if the fan was noisy.

Meanwhile, the main motor should be lubricated with oil every six months. There is a small grub screw next to the motor shaft which needs to be removed. Then six drops of light oil added whilst turning the shaft, before replacing the screw.

If oiling does not improve matters then the only remedy is to fit a new motor. smile


If you don't inspect ... don't expect.
Joined: Jan 2007
Posts: 3
Newbie
Offline
Newbie
Joined: Jan 2007
Posts: 3
I have known an uneven build up af dust on the fan blades to cause a rumble due to being out of balance.

Joined: Feb 2004
Posts: 14,807
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,807
Likes: 72

Dust? What about PM? smile

Yes ... you encounter similar "problems" in grotty PC's.

Even one of my old PC's (1996) squeals a bit (especially when it's a bit cold - as it often is in my garret). I have to let it boot up then stall the fan (motor) - then release it again, for "silence" to return. I don't mind the noise myself, but my neighbours may not enjoy it quite so much!

All of the above is a bit difficult in the Dräger 8000, of course. smile


If you don't inspect ... don't expect.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (Neil Porter, Huw), 1,420 guests, and 24 robots.
Key: Admin, Global Mod, Mod
Newest Members
Nick.A, Tdayve, AdrianT82, Iagonuns, Dctm
10,371 Registered Users
Forum Statistics
Forums26
Topics11,261
Posts74,525
Members10,371
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5