The manufacturer states the device is Class 1 CF.
The safety of the device is not reliant on it having an earth if there are no accessible earthed or uninsulated conductive parts?
Its function might require an internal earth but we are discussing safety testing.
Concerning medical equipment, with applied parts, I think it does. What about applied parts leakage w.r.t earth?, for example, (irrepsective of whether there are accessible earthed parts on the case or not).
How do we know that applied parts on a Class 1 device with "Class 2 parts" are isolated from earth (or with mains on applied parts w.r.t earth) if we test Class 1 devices with "Class 2" parts as Class 2 overall?
Even if we test Class 2 and assume a "functional earth" there's still a requirement to test functional earth leakage and touch currents, etc, plus applied parts w.r.t earth on Class 2 devices is there not?
If we don't test Class 1 then this implies that there's no earth bond test performed on the 3-pin removeable IEC leads fitted to devices with "Class 2 parts", i.e. inaccessible earths.
In my opinion it's better practice to test Class 1 devices (with inaccessible but protective earths) as Class 1, mark the earth as inaccessible, then test the mains IEC earth bond seperately, followed by the tests actually required for a Class 1 device, rather than miss earth-referenced leakage tests, etc, because Class 2 EST is performed.
We must trust manufacturers and notified bodies to classify their products and then test them appropriately - precisely because we're not type-testing. Mains electrical devices are either Class 1 or Class 2 - nothing in-between as far as I'm aware.