If it's being used in Physio ... there are
Dark Side companies (plus one at least who is a major supplier of physio kit) who will be happy to come and PM everything in the department at "very reasonable" rates (in one case that I came across - at less than the in-house NHS biomeds could do it for)!

@KM: "not a medical device"? Surely
all "exercise" equipment used in Physio is like that? But (to my old-fashioned / traditional biomed mind), that shouldn't mean that it doesn't get looked at. In fact, I would argue that it makes it
more relevant to the PM programme (or whatever you want to call it). Bottom line:- if it's used in a "patient" environment [tick], then it gets PM'd [tick] ... and if it's got a mains cable [tick] then it gets an electrical safety test [tick]. There you are, all the boxes ticked!
Lots of junk gets bought by the medics (with good intentions, presumably). That has been the traditional for as long as I can remember. But that doesn't mean that we should wash our hands of it. Rather the reverse, in fact:- the crappier the kit, the more maintenance is likely to be needed!