#39344 - 03/07/09 11:35 AM Re: Renal Dialysis water treatment plant [Re: mtalam]
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1499
Loc: Jeddah, Saudi Arabia
Same as UV sterilization, to be avoided
_________________________
Stress is for other people

Top
#39359 - 03/07/09 05:12 PM Re: Renal Dialysis water treatment plant [Re: Neil Porter]
techman Offline
Expert

Registered: 03/11/03
Posts: 136
Loc: London
Some valid points have been made so far. I would add that you should work to a minimum output. I would suggest you base it on:

Maximum demand from the dialysis machine x number of machines x 1.25

So for 10 machines I would be looking at a minimum output of 16 L/min. If there is likely to be further expansion in the near future, it will be worth factoring that in.

The other point not discussed is the quality of water you want. Are you looking at AAMI or higher quality? Also are you looking to have a back-up system? Both of these aspects will make a sizeable difference in cost.

Top
#39360 - 03/07/09 06:24 PM Re: Renal Dialysis water treatment plant [Re: techman]
Neil Porter Online   content
Hero

Registered: 23/02/09
Posts: 1499
Loc: Jeddah, Saudi Arabia
Most systems have redundancy built in, plus they are upgradeable in terms of output.
_________________________
Stress is for other people

Top
#44426 - 24/02/10 08:02 PM Re: Renal Dialysis water treatment plant [Re: Geoff Hannis]
rug Offline
Technologist

Registered: 23/02/10
Posts: 46
Loc: Canada
Water systems for hemodialysis application can work with any model hemodialysis machine. I have consulted on about a dozen clinic installations/new construction in the last 10 years. Mr Hannis said it well. Listen to him.
I have Gambro and Marcor(US company) water systems and both have proven to be vary reliable. Biomed is responsible for the water systems and hemodialysis machines. Had to learn a lot of plumbing skills. Both offer heat disinfectable systems which work very well and cuts way down on using those nasty disinfection chemicals.
In Canada a water system for hemodialysis application must have a medical device liscence designation. Is this required as well in the UK?

Top
Page 3 of 3 < 1 2 3



Moderator:  DaveC in Oz, KM, RoJo 
Sponsors
Press Releases
Who's Online
12 registered (Laiq, infinix, GeorgeK, Lee S, Huw, ferhat__aydin, bosch, Geoff Hannis, Angus, NickM, DaveC in Oz, Neil Porter), 281 Guests and 8 Spiders online.
Key: Admin, Global Mod, Mod
May
Su M Tu W Th F Sa
1 2 3 4 5
6 7 8 9 10 11 12
13 14 15 16 17 18 19
20 21 22 23 24 25 26
27 28 29 30 31