As an aside: Years ago I once got asked by a manager at interview, in a hospital at Preston, what the voltage was on an energy storage capacitor in a LP9 when the output energy was charged up to 360J. I responded by telling him that it was difficult to give a figure with any accuracy because:
1) There are losses in the waveshaping components R & L thus the energy required in the capacitor to deliver 360J would be greater than expected on the face of it hence this would mean the capacitor terminal voltage would differ from the nominal voltage that's expected with an output of 360J, even if the nominal capacitance value is known.
2) Since the capacitance value of each capacitor is subject to wide tolerances only an estimate of voltage in the range X to Y KVolts can be given (also considering 1 above). I gave a "ballpark figure" but he wanted to know exactly (I guess this person must have had the P.D. on a LP9 unit in the workshop measured at 360J or something).
I indicated I knew the nominal capacitance value in the LP9 (as I'd had to change one or two) and gave the "ballbark figure" for the P.D. across capacitor, charged to around 360J, but he wasn't pleased with my answers I could tell. He then came out with an exact figure that I should have known, apparently. Needless to say I didn't get the job.
So much for the stuff I learned at "O" and ONC level - Talk about a "closed question", eh?