We do the PPM and service for Graseby Syringe pump in our equipment library. We use Metron infusion Pump Tester to test Occulsion level.
I find out that most the time, the 3100 occulsion pressure is below 476mmHg (9.20 psi) (Suggest level from Manual). Normally, they are around 7 to 8 psi before I service them. Therefore, I have to adjust them to 476 – 556 mmHg (9.20 – 10.75 psi) according to the manul.
However, recently, we brought few brand new Graseby 3100 pumps from Smith medical. I checked them, their occulsion level are below 9.20psi.
Can any one who work on Graseby 3000 pumps gives some advices here? Do I make some mistake here or you also experience the same problem?
Many thanks.