Hello David
As previously mentioned by Aston, the interconnect cable on these pumps can also be a problem.
The interconnect cable , in my experience does not last that long. (the one above the transformer is the one i refer to)
Alaris or whatever they are called where you are, supply replacecements, and the new ones are more better than the old ones. They fit better and the creases are in the right place as to avoid the heat from the transformer as much as possible. I think they cost a few pounds each, but in my experience of looking after these pumps, the new connector was superior.
A very common fault with the connector is an error-onous low battery alarm.
I found that with the case open problems with the interconnect cable could be easily reproduced by moving the cable.
The engineers at alaris in uk are a very helpful bunch so ask them!!!
Tracy