We've been using some 50 of these little pumps for 2 years now in several departments -flawlessly.
Have recently introduced them into a Pediatrics
department - knowing they've got the Micro option - but alas there's a problem.
In many cases we keep getting the "hi pressure"
alarm again & again without apparent reason. In
one case the parents had to keep resetting the
pump all through the treatment - very unpleasant.
The nurses used to work with the 597 pumps, and
have never encountered this phenomenon with those.
All pumps occlusion setting was verified to "nor"
which is the middle setting. The Venflon used is
the BD 24GA as needed for children (unlike the 18
GA used in the other departments). This needle was
used with the 597 as well without incident.
I'm hesitant to set the occlusion level to "high"
as this might have repercussions, especially with
small patients.
Could anyone suggest an avenue of approach to this
problem ?
Thanks
David