One problem of using a high current for testing the protective conductors on medical devices is the danger of making a temporary repair. That is, if measured at a low current, the device may exhibit a high PE resistance value (even be almost open circuit) but when tested with a high current will exhibit a low resistance. Then when put back in service (well it did pass) the oxidisation or corrosion of the mains connector (be it the IEC or mains plug) causes the earth to become high resistance again! This can mean the total leakage current of the device may end up entering the patient! (As could be the case in a dialysis machine since it has a type B applied part.)

Maybe the use of a high PE test current should be banned as this may lead to 'real' faults being masked?

If the UK made more use of equipotential bonding leads maybe it would not matter as much?