The calibration is done automatically during the pre-use check. If you need to do any further calibration a field service card is required to access the service menu.
If your Servo-i is failing it pre-use check on pressure transducers I would suggest trying with another expiratory cassette at first and then removing and cleaning the inspiratory valve.