Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 2 of 5 1 2 3 4 5
#13351 16/09/03 7:15 PM
Joined: Oct 2002
Posts: 130
EH Offline
Expert
Offline
Expert
Joined: Oct 2002
Posts: 130
Consider yourself smacked (cheeky whippersnapper wink ) I did not design the bloody things, however, i agree they should be CF but
their not...

Quote:
14. Safety
The machine complies with IEC 601-1 and is protected against electrical shock according to classification B.

The machine complies with IEC 601-2-16: Particular requirements for hemodialysis equipment.

The machine complies with IEC 601-1-2 and EN 5501 1 regarding radio interference.

0086 This product is CE-marked in accordance with the requirements in EC Council Directive 93/42/EEC of 14 June 1993 concerning medical devices.
Courtesy of gambro spec.

Seymour


A problem is a gift to be unwrapped :rolleyes:
#13352 17/09/03 8:22 AM
A
Anonymous
Unregistered
Anonymous
Unregistered
A
Do you use isolation transformers with these devices then?

#13353 17/09/03 8:41 AM
Joined: Oct 2002
Posts: 130
EH Offline
Expert
Offline
Expert
Joined: Oct 2002
Posts: 130
RL, you confuse me, why would you need isolation transformers? Millions of dollars where probably spent during the R&D of these devices, Gambro is a highly respected manufacturer of excellent machines, if they tell me it is protected against electrical shock according to classification B then who am I to argue. confused

Seymour


A problem is a gift to be unwrapped :rolleyes:
#13354 17/09/03 9:00 AM
A
Anonymous
Unregistered
Anonymous
Unregistered
A
Class1B means the device passes the safety requirements for 1B and that makes me very happy for the company but I like to understand why sometimes. Yes of course Gambro are well known and respected Seymour but your statement made earlier about Class1B being totally isolated i.e. floating was incorrect; though I'm not being critcal, just wondering how the classification of a device like a renal dialysis machine can be class1B when the pressure-measurement devices (CVP) used with the renal kit are necessarily classified CF.

Just thinking about isolation regarding Renal kit especially; considering the transmission of fluids, internal to the machine, delivered to/from the patient and potential for leaks, etc. Considerations that have been commented on in this thread by others. Mustn't be a problem, otherwise we would hear about it, of course. Just my curious nature, I suppose....I will download the relevant standards from BSI at lunchtime, I think. Ta.

#13355 17/09/03 9:47 AM
Joined: Oct 2002
Posts: 130
EH Offline
Expert
Offline
Expert
Joined: Oct 2002
Posts: 130
Quote:
In ideal i.e. fault free conditions the answer is No, the dialysate fluid and blood circuit are totally isolated
* the dialysate fluid and blood circuit are totally isolated

I will say it one more time so I am sure you have read it correctly

*the dialysate fluid and blood circuit are totally isolated

I refer to the machine's classification ... AGAIN I WOULD LIKE TO REMIND YOU SIR

Quote:
14. Safety
The machine complies with IEC 601-1 and is protected against electrical shock according to classification B.

The machine complies with IEC 601-2-16: Particular requirements for hemodialysis equipment.

The machine complies with IEC 601-1-2 and EN 5501 1 regarding radio interference.

0086 This product is CE-marked in accordance with the requirements in EC Council Directive 93/42/EEC of 14 June 1993 concerning medical devices.
My statement made earlier about Class1B being totally isolated confused Where did I say That..

If you want to debate my friend then I suggest you take the time to read my posts correctly instead of putting words in my mouth. Goodbye

Seymour :p


A problem is a gift to be unwrapped :rolleyes:
#13356 17/09/03 10:04 AM
A
Anonymous
Unregistered
Anonymous
Unregistered
A
Actual Quote I was referring to was:

Quote:
the dialysate fluid and blood circuit are totally isolated i.e. it is a class 1B device
Initially I thought you may have meant 1BF but and made a typo error - after your comments later, I think maybe not. Read the post where you made comments about the heater being in contact with the fluid - I'm really not getting at you for a possible slip of the keyboard Seymour.

Just wondering about safety issues of kit I'm not familiar with. As for isolation of the dialysate and blood, I was talking electrical isolation. Even with my limited exposure I can appreciate the concern about physical/mechanical/electrical isolation of dialysate from blood. This is also evident in the standards you refer to yourself. I still don't understand why 1B - how old are the machines you refer to? What are the classifications of the latest generation of machines?

#13357 17/09/03 10:08 AM
Joined: Jul 2002
Posts: 123
Savant
Offline
Savant
Joined: Jul 2002
Posts: 123
RL your not alone, this topic has undoubtedly got a few of us exercising the old-grey-matter.

One explanation may be that EN 60601-1-1:2001 Page 22 Table BBB.201 “Combinations of medical and non-medical equipment” requires an additional protective earth or separating transformer where (inside the patient environment) an item of medical and non-medical are in use (the bed or chair being non-medical?). This is also a requirement in the New Draft International Standard IEC6061-1, third edition, which has been out for consultation. The new draft further highlights the hazards of “Touch currents”. Dialysis companies may be reacting to this?

Out of interest, the New Draft further implies an additional protective earth on items of medical equipment (inside the patient environment) supplied via a multiple socket-outlet.

Bill


Bill
#13358 17/09/03 10:19 AM
A
Anonymous
Unregistered
Anonymous
Unregistered
A
Thanks for that WOM - I hope you agree that 1B is not totally isolated (electrically) not as far as I'm concerned anyway.

As you say; combinations of electrical separation through isolation transformers and earthing, when absolutely necessary, are recommended I think. Yes, touch voltages need to be reduced to within the limits, so I suppose the best way is to "tie everything and anything down". What's the standard for electrical safety of chairs? wink

As John S says, the earthing of components of a system should be sufficient - it comes down to isolation really doesn't it? Especially for a single device i.e dialysis machine?

#13359 17/09/03 10:39 AM
Joined: Sep 2002
Posts: 4
Gareth Offline OP
Newbie
OP Offline
Newbie
Joined: Sep 2002
Posts: 4
Thanks all. I've forwarded to Jim Lefever at MHRA for comments. I've got the Fresenius doc as a JPEG but am not allowed to attach to this post. I'll send to anyone for comments.

Gareth

gareth.murcutt@royalfree.nhs.uk

#13360 17/09/03 11:03 AM
Joined: Oct 2002
Posts: 130
EH Offline
Expert
Offline
Expert
Joined: Oct 2002
Posts: 130
Now I am really confused, the original question as posed by Graham was
Quote:
It is easy to see how the fluid is in contact with the patient but how much is the fluid in contact with the machine? Does the fluid actually contact the machine or does it stay within the (I assume insulating) set
And I answered with regards to the "electrical" isolation of the blood and dialysate fluid which are two seperate fluid ciruits connected via a dialyser. What I was trying to say was the Blood and Dialysate FLUID circuits, and the patient are "Electrically Isolated from the dialysis machine".

Seymour confused rolleyes


A problem is a gift to be unwrapped :rolleyes:
Page 2 of 5 1 2 3 4 5

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 10,968 guests, and 71 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5