Has anybody evidence of equipment failing Electrical Safety Testing only in Reverse Polarity. Is Routine Testing in Reverse Polarity really necessary?
I'm no expert at safety testing so please feel free to pick this simplistic reasoning to bits but I think about it this way: The aim of routine safety testing, on acceptance, post-repair, whenever, is to maintain patient/operator safety under normal operating conditions and in the event of a likely single fault condition (SFC) that may occur.
For Class 1 medical equipment (with protective earth) under normal conditions; in the event that there is a degradation in insulation between L (at mains potential) and E (close to Neutral potential under normal conditions - remember Earth is bonded to Neutral at the substation) then unacceptably high earth leakage currents will flow. Complete brakdown L-E will result in the equipment protective fuse(s) in the L circuit blowing under normal conditions.
If the equipment wiring is reversed e.g. somebody reverses the L/N connections in the IUT - a SFC exists and the Neutral circuit in the equipment is raised to mains potential (it usually never achieves mains potential under normal conditions to do this the safety tester seperates N from E and reverses L/N). If there is excessive earth leakage under these conditions then there is degradation in the N insulation to Earth. In addition, under these L/N reverse conditions, if there is a short-circuit between N and E then the N fuse (required in medical equipment) blows.
After manufacture or repair the mains parts of the instrument L/N can be accidentally reversed. Reverse polarity testing ensures the equipment earth leakage is acceptable under likely single fault conditions and safety tests may allow us to identify the source of the problem.
I have to agree with Sean; under SFC conditions where one supply conductor is O/C then if earth leakage increases then the E is completing a return path indicating poor seperation of L or N to earth, within the instrument, depending upon which conductor is O/C and polarity (of L/N). Under normal conditions N is close to E potential so degradation in N-E insulation would necessarily not be detected by testing under normal conditions would it? I will reason it out, later-on, below.
Another point is that safety testing should test for all combinations of SFC being considered; we are not just considering S/C but degradation over time up to the acceptable limits - systematic testing is a means to record values of combinations of the necessary tests laid down in the relevant standards for routine testing.
If the wiring in the mains supply (socket) is reversed i.e. L to N then is it not likely there will be a bang and the mains protection will kick-in? This is due, effectively, to the connection of L to E since N is close to E potential due to bonding of L-N at the power distribution source.
Are you using a medical electrical safety tester to BS EN 60601-1 Lindsay?
I might be missing somthing, but in normal condition testing, with a neutral to Earth short, the ammeter (in place of the earth) will measure the majority of the load current of the equipment. The fault would be discovered before advancing to SFC testing, and an open Neutral.
Let's assume, in this case, the N is effectively in parallel with an E (right back to the power distribution source) with an ammeter shunt resistance, R, in series with E in the safety tester (considering pure resistances only) and the S/C is instrument N to instrument E enclosure.
Total current = Ncurrent + Ecurrent. I suppose the majority of the load current will take the path of least resistance.....2/3rds of the current will flow down N: if N = R and E = 2R i.e. if the N and E are the same length and ammeter R is same as N and E path resistance (not unrealistic). Hence only 1/3rd of all S/C current flows in the E under these, albeit theroteical, conditions; therefore earth leakage would not equal load current by a long chalk so to speak.
Under these circumsatnces; for low power equipment (drawing 0.75A for example) then under normal mains polarity conditions but with N-E S/C Ncurrent would be 500mA and Ecurrent would be 250mA - lower than the maximum acceptable limit permitted. Normal testing only would not pick up the N-E fault. I would be interested to hear others comments on this.