Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 2 of 2 1 2
Joined: Sep 2002
Posts: 139
Expert
Offline
Expert
Joined: Sep 2002
Posts: 139
We have 3300's using a stat of 200ml/hr and using the free flow device with no probs. It just sem to be a problem with the 3200's. Units are adjusted to the factory set value.

There seems to be a bit of ambiguity regarding occlusion levels and apporpriately set prssure alarm levels. Some pundits seem to advocate setting pressure levels just above line pressure in order to have an rapid time to occlusion, others seem to advocate that it doesn't matter. Am I missing something?


Time is of the essence. Don't abuse it. Just make the most of it.
Joined: Sep 2002
Posts: 54
Scholar
Offline
Scholar
Joined: Sep 2002
Posts: 54
Richard, Can you give me the references you are referring to. We to have a similar issue with 3200's.

A
Anonymous
Unregistered
Anonymous
Unregistered
A
Sure - I will dig them out of the in-tray on my bench.

A
Anonymous
Unregistered
Anonymous
Unregistered
A
Yes Joe - the point I made was that the 3200 occlusion specification, if the same as the 3100 (3.5Kgf- 4.2Kgf from memory), may be lower than the 3300 occlusion specification.

Unfortunately I haven't looked at a 3200 manual for years so I'm only guessing (perhaps you could tell me what the specification is and how you check compliance with it).

Anyhow what I was getting at is that if we (and others) have had problems with a similar line/syringe setup using the 3300 (200ml/hr stat rate) then problems with a 3200 at 200ml/hr using manufacturers specifications, for the occlusion threshold, are likely.

Joined: Mar 2001
Posts: 182
Mentor
Offline
Mentor
Joined: Mar 2001
Posts: 182
I was interested in the comments regarding tissuing and occlusion pressures.

I am sure in my own mind that tissuing will occur regardless of occlusion pressure setting if the clinical conditions exist, i.e. cannula movement, fragile vein, poor positioning etc. There are examples of tissuing occuring with gravity infusions where the in line pressures are marginally above the venous pressure to ensure positive infusate flow.

The purpose of occlusion limits as has been well descibed earlier it to prevent bolus delivery and indeed pump or disposable damage, not tissuing. The idea of trending is Ok so long as someone is there to view and interpret the data.

I know many a sales pitch has been built up on the benefits of pressure monitoring and tissuing incident reduction but I remain sceptical. I am firmly in the camp of "routine observation of the IV site" as a method of prevention. If the reply comes back that there arn't enough staff to undertake the observations, then there arn't enough staff to interpret pressure trend data either!

We have remained with the manufacturer's default setting for all infusion devices apart from those for neonates for whom we reduce the setting by 50%. Tissuing events as reported to me have been low to non existant.

Lucky? - possibly but I wait to be convinced.


Brian


**** Still grumbling in the Pennines ****
A
Anonymous
Unregistered
Anonymous
Unregistered
A
I do prefer calibration of occlusion alarm thresholds, externally, rather than the tedious internal, mechanical, adjustments necessary on older generations of syringe drivers - whatever the source of alarms.

A
Anonymous
Unregistered
Anonymous
Unregistered
A
The other potential benefit of pressure monitoring/trending is that some syringe drivers use "fast start" and "backoff" features that remove actuator backlash (which causes initial loss of delivery at startup) by monitoring syringe actuator/line back-pressure and can actually remove the bolus, due to occlusion or excessive pressures in the distended line, by withdrawing ("backoff") the syringe plunger until pressures are in the normal range i.e. on alarm excessive pressure is removed from the system and bolus prevented from being delivered (if the occlusion is subsequently cleared under pressure). The pressure monitoring effectively completes a control-loop.

Joined: Jul 2004
Posts: 9
Newbie
Offline
Newbie
Joined: Jul 2004
Posts: 9
Brian correctly points out that there should be regular monitoring of infusion sites. However, once the operator observes colour change and swelling, damage has already occurred. In neonates particularly it is important to use devices that can measure small changes in pressure. The line pressure will vary depending on syringe size, flowrate and drug viscosity. It is therefore important to measure the line pressure for a few minutes post setup and then set a pressure window on top of that. Should a pressure increase occur, then it will be detected early before there is visible signs of tissue damage. Some syringe drivers allow this accuracy by using pressure monitoring sets. Some users choose to follow this procedure - some do not. If there is an incident and the technology was available and not used, would this be criticised? Is it not true that in the best interests of patient safety particularly in our most vulnerable patients (neonates) that pressure monitoring be carefully applied.

Page 2 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 17,629 guests, and 20 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,248
Posts74,481
Members10,358
Most Online49,431
Yesterday at 12:24 AM
Powered by UBB.threads™ PHP Forum Software 7.7.5