Adrian,
I find that sometimes it’s good to draw a simple circuit diagram of problems like this to help work out what is going on. If anyone would like to send there fax number I could fax you what I would draw. If you do draw your own diagram I sure you will see a link created by the patient, however this link can not be used as an alternative return path. i.e. current leaving Generator A can not return to Generator A via Generator B.
It's only my opinions I'm expressing - I don't purport to be an expert but me not being an expert doesn't make your understanding correct or any more valid than mine. I've not said that the currents don't return to the generator what I've been consistent with is that the path the current takes to return is not necessarily as one would expect - that's why I use quotation marks when describing 'crossed' currents. The implication being that they're not really 'crossing' and not necessarily returning to the other electrosurgery generator and that it's a description used by others i.e Rojo in this case. Saying that I'm not discounting where the HF current flows because it's a complex issue. I have said:
I think that the 'crossing of currents', from single channel units that Rojo refers to, means that a proportion or all the RF currents from each active would be returned via the the other unit's neutral plate since the neutrals would be common RF returns both connnected to the patient. I suppose the RF active currents would have to be compatible i.e. two similar single channel units being used.
Commenting on Rojo's suggestions about the route of the current (via meaning 'by way of') not that I believe the current won't be returned to it's respective generator; I continue:
Return currents in each circuit would have to be within acceptible tolerances (similar to active currents) but 'crossed current paths' might cause unexpected or less predictable return paths that might pass through tissues sensitive to heating that need to be avoided.
During dual operation, using floating output devices, I believe you can get currents flowing between the neutral electrodes - currents effectively 'crossing' between plates - since the active outputs are not synchronised.
The floating nature of the non-synchronised outputs means that the instantaneous voltage on each return plate can be different causing currents to flow between the plates (and I assume between actives and 'all over the place' - depending upon impedances and the tissue beteen both plates and actives).
The likelihood of this depends upon how close the neutral plates are located relative to the surgery site and their respective active electrodes and to each other, the tissue and the impedances of the different tissues between the electrodes and neutral plates.
Current flowing between the neutral plates ('crossing'). Why not if you have generator outputs interacting even if the outputs have different phase and amplitude? And to try to explain what I'm getting at further:
Simplifying things; I think Kirchoff's law still holds for HF current within a bulk conducting volume (patient) - this law still has to be satified whatever the circuit - most of the current required to satisfy this must return to the respective neutral, it's just the route that is unexpected.
In any case when you have two generators giving 'compatible' outputs with a common reference and tied outputs, even if they are out of phase and of differing amplitude, as you've commented yourself, then there is bound to be some level of interaction with currents flowing between areas of differing potentials. We cannot predict with certainty where the currents are going so we ensure that the plates are positioned in areas that should avoid the potential problems I've mentioned.
Adrian in my opinion it is not possible to draw a 'representative circuit' to describe what's happening to currents and voltages that are out of phase and of different amplitudes when connected to the patient and interacting. For a start you can't relate this to electrosurgery output settings and the relative values of impedances between all of the electrodes. The patient is not a uniform, homogenous, conducting volume. The circuit topology is not predictable: the patient varies considerably in shape. You don't know where the plates and active electrodes are going to be applied, etc, etc. But you do accept there is interaction (mixing as you put it, between the active outputs) so that's something.
I think that it's just accepted that the safest option to prevent unpredictable current paths is to assess the best location each neutral plate with respect to the active it's associated with and not try to prove these unpredictable currents don't exist when in practice manufacturers comment on them and are aware of potential problems.
And RICK I think plates must not be allowed to connect for at least four reasons in my opinion:
1). Touching plates might overlap and not contact the patient over it's surface this could present a relatively poor contact area and higher current density at the overlapping return plate.
2). You could get capacitive coupling, at HF, between the plates (see 4).
3). Currents flowing between plate edges, at different potentials, may cause heating and burns over this small area.
4). That if plates touched then the two electrosurgery units would remain at the same potential i.e. they would not float independently of each other (see1).