#6087 11/01/07 3:05 PM
Joined: Jul 2004
Posts: 20
David R Offline OP
Novice
OP Offline
Novice
Joined: Jul 2004
Posts: 20
For information to all out there maintaining Cosmedico UV light systems.

It seems that Cosmedico-Medizintechnik GmbH based in Villingen-Schwenningen are no more. The UK agent Scott Medical are equally shafted so please don't shout at them - it's clearly not their fault.

This company has until very recently supplied UV therapy and SAD light sources to the UK.

They have been purchased by the biggest fish in this field - Waldmann Lichtechnik GmbH, and Waldmann are discontinuing the GP series of UV treatment cabins. I don't know about the extremity treatment systems, but it seems unlikely that these would be kept on too.

It seems that a seperate entity has been set up to support current users. (I have spoken to the company at the address below and the info. above is correct)

They can be found at :

Cosmedico Medical System
Bismarkstrasse 64
78647 Trossingen

+49 (0) 7425 337090

If anyone has any more info. on this or would like to informally share experiences/service tips, now would seem to be a good time to start talking.

Best regards

Dave

Joined: Apr 2007
Posts: 2
Newbie
Offline
Newbie
Joined: Apr 2007
Posts: 2
yes, it makes old tech manuals of gold value.
I m looking for the error codes of a medical Cosmedico GP-42K.
does anyone have it ?
Thanks
EP


Joined: Jul 2004
Posts: 20
David R Offline OP
Novice
OP Offline
Novice
Joined: Jul 2004
Posts: 20
This relates to GP36 and GP42 (all variants).

Error codes F21 to F26 relate to faults in either the relays on the control board (K1 to K6), the lamp group contactors (Confusingly, also referred to as K1 to K3), or the cabling between. There is a diagnostic plug to help narrow the fault further.

The relays on the control board are not designed to be replaced, it is a control-board swap-out in this case.

Dave

Joined: Apr 2007
Posts: 2
Newbie
Offline
Newbie
Joined: Apr 2007
Posts: 2
Thank you very much for your help
EP


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 1,437 guests, and 110 robots.
Key: Admin, Global Mod, Mod
Newest Members
Nick.A, Tdayve, AdrianT82, Iagonuns, Dctm
10,371 Registered Users
Forum Statistics
Forums26
Topics11,262
Posts74,527
Members10,371
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5