|
|
Joined: Jan 2002
Posts: 21
Dreamer
|
OP
Dreamer
Joined: Jan 2002
Posts: 21 |
Hi, We are looking to obtain images (e.g from a camera stack) and store these on a laptop using a composite PCMCIA card. Is there a reason / standard etc that I should be aware of before commencing this? Thanks, John
|
|
|
|
|
|
Anonymous
Unregistered
|
|
Anonymous
Unregistered
|
Hi John,
What are the interconnections between the PC I/O ports and the medical system and will the PC be connected to mains during use? If so how/where are the connections made? What does the stack consist of, i.e. what're its component parts, and how's it all connected?
Richard
|
|
|
|
|
Joined: Aug 2004
Posts: 237
Master
|
Master
Joined: Aug 2004
Posts: 237 |
If it is connected as part of a video stack 'system' it will need to comply with BS EN 60601-1-1. (I think?)
BS EN 60601-1-1:2001 Medical electrical equipment. General requirements for safety. Collateral standard. Safety requirements for medical electrical systems. Section 1.1 Collateral standard: Safety requirements for medical electrical systems
John
There are things that are known and there are things that are unknown. In-between there are doors.
|
|
|
|
|
Joined: Sep 2002
Posts: 139
Expert
|
Expert
Joined: Sep 2002
Posts: 139 |
BS EN 60601-1-1:2001 has been superceded by BS EN 60601-1-1:2006
Time is of the essence. Don't abuse it. Just make the most of it.
|
|
|
|
|
|
Anonymous
Unregistered
|
|
Anonymous
Unregistered
|
Basic aspects of BS EN 60601-1:2006 and BS EN 60601-1-1:2001 are a start. The PC will have to be compliant with its own safety standard, e.g. BS EN 60950-1:2006, and be CE marked if it's connected as part of a medical system via a functional connection, e.g. a composite video cable, and/or a shared mains supply, e.g. MPSO connected to the secondary winding of a separating transformer.
I'm not going to assume you are using a video stack that's compliant with 60601-1 or 60601-1-1, i.e. in terms of safety requirements. Depending upon whether the PC is used on the video stack in the patient environment, in the same room or another room depends upon what's required to connect it to the video-endoscopy system. If you have more details about it then I'm willing to help with more practical advice off forum if you like.
I'm not assuming anything, that's why I asked about the system connections, etc. The standards don't really tell you much unless you have some idea what you're trying to achieve in terms of electrical safety. YOU become the manufacturer of a medical system when you connect devices together - there are higher risks involved when connecting medical and non-medical devices together or you take advice that's flawed.
Alternatively there are always experts at the MHRA that you can speak to if you wish to find out what the safety requirements or implications of adding non-medical devices to medical systems are.
Richard
|
|
|
|
|
Joined: Feb 2004
Posts: 14,808 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,808 Likes: 72 |
Surely it depends if there is already some sort of storage device on the stack (DVD recorder etc.). If so, then the saving to the laptop can be done later, out of harm's way. What's the actual set-up there, John? 
Last edited by Geoff Hannis; 04/06/07 8:28 PM. Reason: Added an apostrophe!
If you don't inspect ... don't expect.
|
|
|
|
|
|
Anonymous
Unregistered
|
|
Anonymous
Unregistered
|
Thats a good point. Alternatively the PC could be used on battery to avoid some problems with providing mains isolation but this isn't ideal, temporary and risky since operators can connect mains, etc. Not recommended.
You'd still need to ensure acceptable leakage currents i.e. "equivalent enclosure and patient leakage" from between conductive accessible parts of the internally powered PC to patient applied parts of the medical system and earth, etc.
|
|
|
|
|
Joined: Jan 2002
Posts: 21
Dreamer
|
OP
Dreamer
Joined: Jan 2002
Posts: 21 |
Thanks so far chaps. The stack is compliant with BS EN 60601-1-1:2006. We want to get away from video printers, etc. The reason for connecting a separate laptop is so we can download the images on to the hospital network as IT have just invested in a terabyte server with a big partition for archiving images.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,808 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,808 Likes: 72 |
No existing storage device on the stack, then? How, and where, is the image collected, and ported to, I wonder (ie, where do you plan to connect to)? The camera controller? What connections are available at the back of that one? Dare I mention Bluetooth, Wi-Fi etc.? That way your laptop could be in the next room! I presume you want to save to the laptop for editing etc., or else why not download directly to the big bad server? This is interesting. Don't forget to let us know how you get (got) on, John. 
If you don't inspect ... don't expect.
|
|
|
|
|
|
Anonymous
Unregistered
|
|
Anonymous
Unregistered
|
If you connect to the network when the stack is in use, via a laptop online, then there is the potential issue of leakage currents being introduced by the network connection(s) under normal or single fault conditions (earth-loops?), irrespective of whether mains separation being used to supply the laptop PSU and the composite data is being acquired from a medical grade video processor.
How things are interconnected as a medical sytem is the issue - not just whether, as individual devices, they meet their respective standards.
Table BBB.201 in 60601-1-1 (Safety Requirements For Medical Systems) and reference to the rationales in the 3rd edition 60601-1 standard is useful when considering the combinations of medical electrical equipment and how they're connected RE: solutions to providing additional protection to medical systems.
Just because each individual interconnected devices meet their respective standards doesn't mean that the system necessarily meets safety requirements.
|
|
|
|
Forums26
Topics11,265
Posts74,543
Members10,376
| |
Most Online59,530 Apr 30th, 2026
|
|
|
|