Quote:
Today the company sent a service engineer to carry out EST the ECG monitor. It failed there own EST. The ECG monitor passed the other parts of the safety test with no problems. The company is running two different IEC 601-1 standards.

Workshop engineers are using a lower standard: 2 Mohms

Field engineers are using higher standard: 50 Mohms


Is it possible that the 2MOhm minimum limit is actually that for L/N-Earth (Class 1 devices) and the 50MOhm limit is for AP-Earth (Class 1, BF/CF)?

Most "'60601-1 compliant" medical grade electrical safety testers in common use (that I'm familiar with anyhow - about 5 types) test L/N-E for all class 1 devices and AP-Earth for class 1 BF or CF if they're indicated.

It's likely, I think, that a bit of confusion as to what the safety testers are measuring may be complicating the issue. Do the field engineers use a different safety tester than the workshop engineers by any chance?

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
Anything more on this one, Aston, or is that it for now? smile


If you don't inspect ... don't expect.
Joined: Sep 2004
Posts: 213
Master
OP Offline
Master
Joined: Sep 2004
Posts: 213
No up dates as of yet.

The company has now bought in older, cheaper patient monitor for trail. It passed it EST first time.

I just find it odd that a company can try and push these monitor on to us so hard.

I would always through IEC601 testing would be part of the CE marking.

I will let you company and product when the trail has finished.

A.M boggle


Barry

Be not afraid of greatness; some are born great, some achieve greatness, and others have greatness thrust upon them
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
Thanks for the update, Aston. We're all waiting in eager anticipation, no doubt about it.

How about Older + Cheaper = Better!

Meanwhile, why not delete odd and insert disgraceful (?) smile


If you don't inspect ... don't expect.
M
Mr R J Ling
Unregistered
Mr R J Ling
Unregistered
M

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
Darn! I thought the CE mark meant Made in China! smile


If you don't inspect ... don't expect.
Joined: Sep 2004
Posts: 213
Master
OP Offline
Master
Joined: Sep 2004
Posts: 213
Just for Geoff.

The piece of equipment was GE Dash 2500.

Now others are also having problem.

GE Dash 2500 - electrical safety

Hope this Helps

A.M boggle


Barry

Be not afraid of greatness; some are born great, some achieve greatness, and others have greatness thrust upon them
Page 2 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 8,505 guests, and 293 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,285
Posts74,589
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5