Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
#31370 16/06/08 7:05 PM
Joined: Jul 2000
Posts: 1,965
Likes: 32
Hero
OP Offline
Hero
Joined: Jul 2000
Posts: 1,965
Likes: 32
We have several non-lumened endoscopes in the trust - nasal endoscopes and flexible laryngoscopes. These are used mostly in OPD areas and also sometimes in theatres. As they are non-lumened they do not need decontamination in the (expensive) endoscope processors. They do however still need effective decontamination between uses. Usual practice is a wipe process which has been in place for ages. However this is not validatable.

Can anyone advise what best practice is for decontamination of these scopes?


Be Proactive and reactive.
Joined: Dec 2003
Posts: 105
BK Offline
Savant
Offline
Savant
Joined: Dec 2003
Posts: 105
John,
Validation is the key word. Manual cleaning/reprocessing cannot be validated and effectiveness is only as good as the person doing the cleaning. We use sheaths for such scopes, they can however work out expensive. There is a small automated table top washer/disinfector available from Dawmed, they did a costing exercise against the use of sheaths which showed some savings, the main advantage would be that you get a printount (validation)for the reprocessing cycle. We did not persue the offer therefore I am unable to give an opinion about the machine. We are currently looking to centralise all endscope reprocessing.

Joined: Jul 2003
Posts: 108
Savant
Offline
Savant
Joined: Jul 2003
Posts: 108
John,
The watchwords are Validation/ Automation (as stated already).

Have a word with your sterilizer 'AP' for specific advice and thoughts on current practices in use. Many of them will advise directly on these issues as will your Microbiologists.

We are considering bench top machines for this type of thing (Nasendoscopes) although costs are high for such things.Current practice is sheaths and manual cleaning techniques.
Turnaround and throughput of scopes in our ENT clinic/s is very high negating any centralisation option.

I'm farly new to a direct (full-on) decontamination remit but have become aware, quite quickly, that there are many important issues to consider in such cases as this.

Overall - i feel that issues around all aspects of Scope decontamination will go the way of SSD processes for surgical intstrumentation - it's a matter of time before central guidance comes out more fully as a way of improving general standards across the board.

Joined: Jul 2000
Posts: 1,965
Likes: 32
Hero
OP Offline
Hero
Joined: Jul 2000
Posts: 1,965
Likes: 32
Thanks BK & Darlo,

I'll look Dawmed up on Google and give them a ring, but it would probably be cheaper to put them through SSD. (If they will do them)


Be Proactive and reactive.
Joined: Jul 2003
Posts: 108
Savant
Offline
Savant
Joined: Jul 2003
Posts: 108
John,
Yes, but whatch your temperatures for any chosen automatic process versus equipment compatibility and keep an eye on the turnaround issue from busy clinics. It's the required throughput in the departments that affects number of scopes required by the users.
Imagine a 'quick wipe down' (1 minute duration?) versus a validated disinfection process (25 mins plus?) for a scope.

You can quickly imagine how this could impact on the patient throughput in the clinic.

Regards - Steve

Joined: Feb 2007
Posts: 6
Newbie
Offline
Newbie
Joined: Feb 2007
Posts: 6
Yes, Steve is right. Flexible endoscopes should not be subjected to temperatures above 50 - 60C (depending on manufacturer) and so are not usually handled by SSD departments. Some trusts are going down the washer disinfector route for ENT outpatients, but are having to double or triple their scope fleets just to maintain lists. In my experience, ENT outpatients see lots of cases with little time in between for washing scopes. www.entuk.org has some good basic disinfection guidelines for this type of endoscope.
Regards, Andy

Joined: Jul 2000
Posts: 1,965
Likes: 32
Hero
OP Offline
Hero
Joined: Jul 2000
Posts: 1,965
Likes: 32
All looks very expensive - whatever i do frown


Be Proactive and reactive.
Joined: Dec 2003
Posts: 282
Master
Offline
Master
Joined: Dec 2003
Posts: 282
Our disinfection committee went down the washer disinfector route in our ASDU. Due to the time of the process we have gone from 1 to 6 scopes. Not cheap. Anybody with Keymed flexible scopes should go to Keymed for advice on the use of Sterilox. The last PPQ I have, dated Jan 2008, states that it known to cause functional damage.

Joined: Feb 2004
Posts: 14,802
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,802
Likes: 72
When you think about it, stuff that powerful is bound to cause some sort of deteriorations! All disinfection chemicals (even the bleach under the kitchen sink) needs to be used with care. To me, that's the real issue right there ... the "quality" of CSD (insert whichever acronym you prefer) staff themselves! I well remember a guy who never bothered to close the main water valve every evening ... then went off on holiday ... a plastic connector, under pressure, "popped-off" ... and no-one phoned us until there was water running into the elevator shaft (and still none of the staff had the nous to simply wade in there and close the stop-valve)! What a mess! smile


If you don't inspect ... don't expect.
Joined: Dec 2003
Posts: 105
BK Offline
Savant
Offline
Savant
Joined: Dec 2003
Posts: 105
Graham,
That is the point I made in "Sterilox" thread.
I would recommend obtaining written confirmation from scope manufacturers that their scopes are compatible with the disinfectant being used

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (Diego Lins, daisizhou), 6,068 guests, and 32 robots.
Key: Admin, Global Mod, Mod
Newest Members
Janet, james_on_ebme, Biomed Egypt, Nektarios K., Keisya Andretta
10,364 Registered Users
Forum Statistics
Forums26
Topics11,254
Posts74,506
Members10,364
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5