Joined: Apr 2005
Posts: 266 Likes: 1
Master
|
OP
Master
Joined: Apr 2005
Posts: 266 Likes: 1 |
Dear friends,
We have a problem with infusion Pump BBraun. We test the equipment with IDA4, BBraun says the equipment the test is wrong. Sameone have experience with that question?
Best regards,
Eugénio Oliveira
|
|
|
|
|
Joined: Jan 2002
Posts: 161
Expert
|
Expert
Joined: Jan 2002
Posts: 161 |
Bom Dia Eugenio!
What kind of Braun Infusion device is it??
Richie.
|
|
|
|
|
Joined: Apr 2005
Posts: 266 Likes: 1
Master
|
OP
Master
Joined: Apr 2005
Posts: 266 Likes: 1 |
Hi,
The model the BBraun is FM, FMS, Compact, Compact S. Braun says the mesures with IDA are wrong...
Regards,
E.O.
|
|
|
|
|
Joined: Jan 2002
Posts: 161
Expert
|
Expert
Joined: Jan 2002
Posts: 161 |
Yes they are. The procedure for servicing of these items is done by use of a special to type syringe calibration device and software supplied from BBraun. You then hook up to your PC and do all the calibrations via that.
I'm not sure that BBraun will supply you with the software and cal syringes unless you have done their engineers course. You can try though.
Otherwise give me a shout and I'll come and do them myself!! I used to live in Cascais, great place!
|
|
|
|
|
Joined: Aug 2005
Posts: 19
Novice
|
Novice
Joined: Aug 2005
Posts: 19 |
Originally posted by Snowler: Yes they are. The procedure for servicing of these items is done by use of a special to type syringe calibration device and software supplied from BBraun. You then hook up to your PC and do all the calibrations via that.
I'm not sure that BBraun will supply you with the software and cal syringes unless you have done their engineers course. You can try though.
Otherwise give me a shout and I'll come and do them myself!! I used to live in Cascais, great place! But Mofe also mentioned fmS and fm which are volumetric infusion pumps - not syringe pumps so you can not use the calibration device.I use Metron QA-IDS which seems to work fine.If there is any problems (malfunction,error indications etc.) I use the syringe driver and software.
|
|
|
|
|
Joined: Jul 2002
Posts: 2,020
Hero
|
Hero
Joined: Jul 2002
Posts: 2,020 |
Mofe, As you are having problems with so many different models of pump can I ask an obvious question: Have you checked the IDA? I have known these to be consistently out, making the pumps look inaccurate, especially if it has anything other than distilled water through it. Robert
My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
|
|
|
|
|
Joined: Mar 2005
Posts: 61
. Scholar
|
. Scholar
Joined: Mar 2005
Posts: 61 |
Braun use the IDA4 FOR FLOW MESUREMENTS AS DO MOST PUMP SUPPLIERS AND IF CALIBRATED YEARLY SHOULD BE CORRECT .THE CAL SYRINGE AND SOFTWARE YOU TALK ABOUT IS USE TO CALIBRATE THE PUMP AS WITH ALL TYPES OF PUMPS I THINK YOU ARE ALL TALKING AT CROSS POINTS .ALSO PEOPLE SHOULD NOT USE NICK NAMES AND STUPID TITLES YOU SHOULD BE PROUD TO USE YOUR REAL NAMES OTHER WISE IT MAKE S THE EBME SITE LOOK LIKE A CHAT ROOM.
|
|
|
|
|
Joined: Aug 2005
Posts: 19
Novice
|
Novice
Joined: Aug 2005
Posts: 19 |
Originally posted by John Harte: THE CAL SYRINGE AND SOFTWARE YOU TALK ABOUT IS USE TO CALIBRATE THE PUMP AS WITH ALL TYPES OF PUMPS I must ask how do you calibrate volumetric infusion pump like B.Braun Infusomat fmS with calibration syringe? I think it's impossible.So you must use Metron QA-IDS/Lagu or IDA 4 in my opinion. According to my knowledge the cal syringe can be used only for Perfusor fm and Perfusor compact-pumps - not for Infusomat fm/fmS-pumps.
|
|
|
|
|
Joined: Mar 2005
Posts: 61
. Scholar
|
. Scholar
Joined: Mar 2005
Posts: 61 |
I use the IDA4 for flow and occlusion test.The cal syringe and software is used to callibrate the syringe drivers (pumps) most syringe pumps use simillar methods to calibrate their syringe pumps
|
|
|
|
|
Joined: Jan 2002
Posts: 161
Expert
|
Expert
Joined: Jan 2002
Posts: 161 |
I was referring to the BBraun perfusor compact when I spoke of the requirement for special to type software and syringe calibrator.
Can we call off the Syringe Police now?
|
|
|
|