If want to check the IDA 4 the best way is to delivery a set flow over a set time and using a very good calibrated weighting scales calculate the acutal volume delivered .But if you IDA 4 IS CALIBRATED BY A CERTIFIED BODY AT LEAST ONCE AYEAR THERE IT SHOULD BE CORRECT

#3717 07/11/05 4:07 PM
Joined: Jun 2000
Posts: 2,412
Likes: 13
Huw Offline
Hero
Offline
Hero
Joined: Jun 2000
Posts: 2,412
Likes: 13
Can we stay on topic please guys?

If anyone would like to re-hash the pseudonym issue - there are existing threads.

#3718 07/11/05 4:24 PM
Joined: Jan 2002
Posts: 161
Expert
Offline
Expert
Joined: Jan 2002
Posts: 161
For Eugenio:

BBraun advised me that using the software & cal syringe is the only way that the compact can be correctly calibrated. There is no reason why you can't use the IDA4 to verify its accuracy, but BBraun recommend that the bi-annual calibration procedure (as descibed before the pseudonym war broke out!) is the only requirement to verify this equipment.

BBraun equipment training is now carried out by Datex Ohmeda, who are based in Hatfield in the UK.

Good luck!

James Bond Lovegod III

#3719 07/11/05 4:42 PM
Joined: Oct 2005
Posts: 49
Technologist
Offline
Technologist
Joined: Oct 2005
Posts: 49
Is that the site or just the thread Rick.

As for that address in Hatfield thats also GE training.

Also where is the NI flag as I come from that neck of the woods.

#3720 08/11/05 11:50 AM
Joined: Apr 2005
Posts: 266
Likes: 1
Mofe Offline OP
Master
OP Offline
Master
Joined: Apr 2005
Posts: 266
Likes: 1
Hi,

Thanks everyone to help me. If i understand is no problem to verify the flow measurements and oclusion pressure in the BBraun infusion with IDA 4.
We have four IDA 4 and they go calibrate every year.When i need calibrate volumetric pump or syringe pump i send the equipment to BBraun, they calibrate and we pay the service, i now it`s wrong, but i don`t have another way. In Portugal we make assistance about 600 volumetric and syringe pump BBraun.
Equipment BBraun is very good, the technicians in Portugal are very good too, and they help me lot of time, but BBraun don`t make training... that`s a pity.
Snowler says BBraun training is carried in Datex Ohmeda. Please give me the adress? Mail?

Sory if i don`t write very well English... I try

Best Regards,

Oliveira Eugénio

Page 3 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 3,652 guests, and 122 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,289
Posts74,595
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5