Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
#37251 19/03/09 1:44 PM
Joined: Sep 2003
Posts: 260
Topper Offline OP
Master
OP Offline
Master
Joined: Sep 2003
Posts: 260
Hi all
Seems that after the conversation i just had with a sales rep, the misguided notion that class 2 equipment requires no safety testing is endemnic. It would seem to me that some people c/o out 'automated'EST without knowledge of what they are actually doing. Surely if Class 2 devices require no EST then why is such facility built into the 601 / Rigel etc. Are such people just assuming all is well internally? i.e. the double insulation is intact?
Tony

Topper #37253 19/03/09 1:58 PM
Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
What about applied parts?
And it you read up the correct way to do the way you test, you will see there is a way to do the eclosure leakage. It is measured with an area of foil pressed against the case representing the palm in case there is capacitive or inductive coupling.
There does seem to be a dangerous attitude that if you cannot clip the test probe on you do not bneed to do the test.
My other concern is with people who do earth continuity tests to the earth stud. What about continuity to the cover you just removed????
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
RoJo #37256 19/03/09 2:13 PM
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

It's all part of a devilish plan to "dumb down" biomed! All automatic electrical safety testers should be withdrawn or disabled!

Surely you haven't missed recent comments on the forum suggesting that some kit doesn't need to be PM'd at all?

People who say that Class II kit doesn't need to be EST'd clearly need to be shown the door (which should be the default action for sales reps, anyway)! Likewise people who talk about a two-year maintenance interval for infusion devices. frown


If you don't inspect ... don't expect.
RoJo #37257 19/03/09 2:14 PM
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
I thought that the main arguement against testing class 2 equipment is that they almost never fail. Most of the electrical problems are easily picked up during a visual inspection.

According to the MRHA (a few years ago) a greater percentage of all problems reported are introduced by equipment being worked on than are found in EST.

having said that we still test.

Lee


Don't forget "we've never had it so good".
Lee S #37258 19/03/09 2:20 PM
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Patient safety (and especially electrical safety) is the raison d'ĂȘtre of biomed!

Anyone who does not sign up to (or, perhaps, has forgotten) this core belief may expect to be visited shortly, in order that they may reap the benefits of a period of "re-training"! wink


If you don't inspect ... don't expect.
Topper #37267 20/03/09 10:52 AM
Joined: Jul 2005
Posts: 15
Novice
Offline
Novice
Joined: Jul 2005
Posts: 15
What type of foil would you recommend? As we have not got any foil would cling film be ok ??

chimps #37268 20/03/09 11:06 AM
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

If you can't scrounge up a bit of "kitchen foil" from one of your mates' sandwich boxes, I have also seen people using a suitably sized piece of sheet aluminium, or even an old diathermy (ESU) patient return plate. smile


If you don't inspect ... don't expect.
Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
Quote:
they almost never fail

That is because people do not test them correctly.
Anyway almost is not never.

I think that most electical failure is either caused by rough handling or incorrect replacing of parts rather than breakdown of insulation due to age.
But when do we do ESTs? After delivery by a courier, being dropped by the user - rough handling - or after we have taken them apart.
If we are doing the test when they are most likely to fail we should be doing them right. As for a yearly tests - well if you have the time to spare and nothing better to do...... shocked
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
RoJo #37274 20/03/09 1:49 PM
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

There's no problem with doing yearly EST. It's part of PM, after all. smile


If you don't inspect ... don't expect.
RoJo #67710 23/05/14 1:03 PM
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Originally Posted By: RoJo

And it you read up the correct way to do the way you test, you will see there is a way to do the eclosure leakage. It is measured with an area of foil pressed against the case representing the palm in case there is capacitive or inductive coupling.


I agree with you, Robert - this has certainly been "custom and practice" for many of us; but I've been trying to find a link (or other source) I can point others to for that one. No luck so far. Any clues? think

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 18,902 guests, and 31 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,249
Posts74,482
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5