#4124 02/02/06 11:50 AM
Joined: Nov 2000
Posts: 71
Likes: 16
Scholar
Offline
Scholar
Joined: Nov 2000
Posts: 71
Likes: 16
Yes we have had the problem too. We have one Trim II that is 5 months old and the lead near the PIM failed. Philips Medical swapped it out under warranty and I also mentioned that the holder for it is a poor design. They said they were looking into changing the design of that in the near future.


Solihull Hospital
University Hospitals Birmingham
#4125 03/02/06 11:23 AM
Joined: Jul 2000
Posts: 969
Likes: 1
Roy Offline
Philosopher
Offline
Philosopher
Joined: Jul 2000
Posts: 969
Likes: 1
If the module just stops working, there's a small satelite board inside which shakes loose when the module is dropped. They are re-designing the whole module to get rid of this extra board - apparently. Hopefully that will improve reliability - and reduce cost of ownership !

Roy


Today is the day you worried about yesterday - and all is well !
Joined: Oct 2004
Posts: 27
Dreamer
Offline
Dreamer
Joined: Oct 2004
Posts: 27
yes we have had several problems with the trim II we have two in A & E, and both have had there PIM module replaced on more than one ocasion the worrying thing is that you have no indication of PIM failure unless you go into service menu. If you follow these instructions you can access service utility screen.. Boot machine from cold and hold down left control and shift keys together and trim knob, you will then be asked for password type 1573 and you will see a window called pim status check there you will see if the lead has failed. and is the only way you can check by swapping with another lead or pim block. the MHRA have been informed of these premature failures of the PIM module and PHilips KNow there is a design fault in the module regarding early failures. they never tested them in a hospital situation with real nurses.

Page 2 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
3 members (daisizhou, ashley, Diego Lins), 8,888 guests, and 21 robots.
Key: Admin, Global Mod, Mod
Newest Members
Janet, james_on_ebme, Biomed Egypt, Nektarios K., Keisya Andretta
10,364 Registered Users
Forum Statistics
Forums26
Topics11,254
Posts74,506
Members10,364
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5