Anyone who has experienced using this equipment(Biotek VT-2)?
I'm choosing between Biotek VT-2 and Timeter 300?
Which do you think is better?
Any help is appreciated.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

I'm also looking for information about the VT-2 (and VT-1). Does anyone have a .pdf manual, and/or any other information? We have found one that is marked VT-2A on the rear of the "display" module, but VT-1 on the bellows arm. Can this be correct? I ask as we cannot (immediately) get the thing to play with the Lung Select lever in the "Infant" position. frown

Furthermore, the short-form instructions on the bellows arm mentions a "Display Hold" button, but there is no such key as far as I can see. Very cunning. I'm beginning to wonder if what we have before us is actually a bastardised unit!


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Still looking ... smile


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

... just taking the Summer one flower at a time! smile


If you don't inspect ... don't expect.
Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
Hello Geoff
(Didnt spot your post till this morning).

Re your points:
As you have correctly identified the VT1 provides testing for Adult Lungs, and the VT2(A) Adult and Infant. The two models share common components but some significant hardware and firmware differences.

The fact that have a lung selector lever indicates a model VT2. The bellows arm (Adult) is also common to both models (irregardless of marking).

On the VT2 there is a mechanical and electrical selection process which must be enabled prior to using Infant mode.

When you indicate "will not play" can you be more specific Geoff ? Are you getting unexpected volume/flow/pressure readings ?

Regards

Sean

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Basically ... the beast doesn't respond when in (what I would call) "Infant Mode". That is with the Lung Select lever to Infant, and the Adult/Infant switch set to "Infant".

When I puff into the thing (in Infant Mode), it feels as though something is blocked, if you know what I mean. smile

Meanwhile, my ISA-Pro problem is still awaiting attention.


If you don't inspect ... don't expect.
Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
Hello Geoff

Been looking into this one.
First thoughts tubing blockage, or too high a volume being injected into the infant lung. A bit too simplistic though. Try following the test protocol to see if you can isolate the problem. As you know yourself, trying to diagnose via a crystal ball is sometimes difficult ....


The settings for simulated normal breathing in a normal patient for VT-1B or VT-2 are as follows:

Adult
Compliance (0.05 L/cmH2O) Resistance (Rp20) Tidal Volume (800 mL) Rate (10BPM) I:E (1.15)

Infant
Compliance (0.003 L/cmH2O) Resistance (Rp50) Tidal Volume (50 mL)
Rate (30 BPM) I:E (1:2.33)

Paediatric
Compliance (0.01 L/cmH2O) Resistance (Rp20) Tidal Volume (250 mL) Rate (20 BPM) I:E (1:2)

If you are obtaining unsatisfactory display readings possible solutions are as follows.

Low or High IE, inspiratory or expiratory times
Perform Full Test, subtract Exp Hold Time from Exp Time, and manually divide this number by Insp. Time to get an I:E ratio reading. If the reading is acceptable, I:E ratio definition used by manufacturer is different than ANSI/ISO version.

High inspiratory flow or high respiration rate
Ensure that the surface on which the Ventilator tester is placed and/or patient tubings are stable and not vibrating or moved during testing. Ensure restrictor and sliding adaptor are properly connected to the airway inlet. Check Calibration, Triggering. If high flow readings persist check Calibration, Volume.

Low inspiratory flow
Ensure proper restrictor is used. Follow instructions for calibration, Volume.
Follow procedure for external flow measurement. If flow readings obtained is acceptable, the ventilator is not compensating for lung back pressure.

Low or High Volume
Follow instructions for Calibration, Triggering.
Ensure compliance selected during set-up corresponds to adult compliance spring or infant compliance lever settings.
Ensure reference units selected in set-up match ventilators reference units.
Press Escape, disconnect patient tubing from VT inlet, move adult compliance spring to 0.15 L/cmH2O position if testing adult ventilator. Repeat test. If volume reading obtained is more acceptable, ventilator is not compensating for lung back pressure. If reading is unacceptable, follow instructions for Calibration, Volume.

Low or High Pressure
Follow instructions to zero transducers. If problem persists refer to Calibration, Triggering and Pressure Checks.

Calibration and Transducer Checks

Triggering mechanism
During any other tests other than Leak/compliance, Ext. Pres., or Ext Flow Tests, observe when you hear a "beep" with respect to the breath cycle phases. If this "beep" occurs at any other time other than just when the Adult lung plate rises during the inspiratory phase, the transducers need to be zeroed due to temperature drift. Follow instructions to Zero the transducers. Normally this should not be required unless the unit has been recently switched on, or has not had sufficient warm up time. Allow at least 30 mins before using any test features.

Hook-up for volume and pressure calibration checks
Disconnect all patient tubing from ventilator ton the VT. Wait until Adult lung plate is horizontal or for 5 seconds (if you were in infant mode).
Connect a 3 way connector between airway pressure tube and the airway pressure port to which the tubing was connected. Connect a 2 way connector to the lung drain tube located in the back and underneath the adult lung. Connect a calibrated pressure meter with +/- 1% accuracy or better and 0.01 cmH2O resolution to each of the connectors. Select Adult positions for Adult/Infant toggle switch and Adult./Infant lung select.
Press Escape key, then Zero key twice, Offset the pressure meters to read zero. Select ATPX for "other volumes and flows" and 0.05 L/cmH2O compliance in set-up, then press escape. Connect a restrictor, the ON/OFF valve and a calibrated syringe of known volume (preferably over 50mL) to the airway inlet/sliding adaptor. Switch ON/OFF valve to ON position.

Volume and Pressure calibration Checks
Inject the known volume into the Adult lung, switch the ON/OFF valve to OFF.
Press volume monitor key twice, volume reading on VT display should match syringe volume injected, if not optical encoder may be faulty.
Press airway pressure monitor key twice, readings displayed by VT should match, if not airway pressure transducer may be faulty.
Press Lung pressure monitor key twice, reading displayed by VT should match the readings on the pressure meter connected to the lung. This reading should be equal to the airway pressure reading obtained above. If not, the differential pressure transducer may be faulty.

Let me know how you get on

Sean Fearon



Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Wow! Thanks for that, Sean.

Should be enough there for me to be going on with!

I'll get to it later.

Cheers

Geoff smile


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

During the start-up sequence:-

Code:
SET-UP Compliance:  *.05   .003
Units in L/cmH2O     .02   .001
                     .01

Well, for Infant Mode it seems that the compliance needs to be set at either .003 or .001 (selected by positioning the * using the [OPTION] key).

Like everything else, it's obvious once you see it!

OK ... so we have the beast working in both Adult and Infant Modes. Later on (when I have time) I'll get a "crib sheet" (aka Idiot's Guide) stuck to it.

Meanwhile, does anyone have a manual (service or users')?

Here's another one for you, Sean. Is there any way to save all those settings entered each time at start-up (is there a battery inside somewhere)? And ... I still haven't found the [DISPLAY HOLD] key mentioned on the istructions listed on the bellows arm.

And (lastly) does anyone have any info about Compliance? Typical values, how to "guess", and stuff like that? smile


If you don't inspect ... don't expect.
Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
Hello Geoff
There is a battery backed RAM chip for the calibration values for volumes etc. which are adjusted by the manufacturer, if this fails the unit firmware provides default values to enable basic operation

Start up requires that you manually set up amongst other parameters atmospheric pressure, relative humidity, gas temperature, as these setup values have a significant effect on VT result's accuracy. A pain but you can see the logic in this.

As for compliance - the general rule of thumb is the lower the numerical value selected for L/cmH2O on either the adult or the infant lung the harder it is to inflate the test lung (remember if you over over pressure you get an audible tone when pressures exceed 125 cmH2O inside the adult or infant lungs). A mid-scale selection is better in Adult mode. In infant (paediatric or neonate) you only get one choice for either.

Regards

Sean

As for the Display hold Geoff, I cant find it either !

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 5,583 guests, and 194 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,290
Posts74,597
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5