As a requirement of ISO audit our 2 601 pro always goes to local Fluke agent for calibration and repair. It already went twice for repair, I remember one of the units always showing high ground leakage even if the it was ok on the other unit. It was repaired by Fluke.
Can you send the unit to your local Fluke agent?

Joined: May 2010
Posts: 153
Likes: 2
michal1 Offline OP
Mentor
OP Offline
Mentor
Joined: May 2010
Posts: 153
Likes: 2
I'm afraid I will send it to Fluke as final solution.
Our local agent will send this device to Netherlands, repair and shipment will be much more than device is worth.
I need to be prepared mentally. Maybe something simple can be done to omit that traumatic experience:)
regards,

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Don't forget that Bong works at a place where funding is no great problem! smile

Just like the majority of folk who get on this forum, he probably also forgets that some of us work on the Dark Side of the Force (that is, small independent operators working in the Public Sector ... and therefore definitely not supported by government funding)!

As you suggest, Michal, we need to explore all options ... which may end up dipping into the savings jar, but certainly also include fixing the things ourselves (or even, as a last resort, binning the thing ... depending upon how much it owes us, as it were).

Perhaps the real message that is emerging here is that the 601-Pro is a great big lump of [censored]! frown


If you don't inspect ... don't expect.
Joined: May 2010
Posts: 153
Likes: 2
michal1 Offline OP
Mentor
OP Offline
Mentor
Joined: May 2010
Posts: 153
Likes: 2
Good news,
I have learned that electrical installation in my flat is old fashion and does not have earth circuit.
That was the problem of mentioned errors. I tried device in newer builiding, and that problems did not occur.
I hope there will not be more issues with the tester, as I can now see it will be difficult to do anything in terms of service with it.
Anyway, I will really, really, really appreciate some service manual......
All the luck for You Guys.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Good news?

Best be careful then when using the iron, microwave oven, vacuum cleaner ... and all the rest! frown

As far as I am aware, the situation you describe would be illegal here in the UK. Better have a word with your landlord, I reckon.

So that way it would be good news ... that the Bio-Tek saved your butt! smile


If you don't inspect ... don't expect.
Joined: May 2010
Posts: 153
Likes: 2
michal1 Offline OP
Mentor
OP Offline
Mentor
Joined: May 2010
Posts: 153
Likes: 2
I am sure it would have been much more things illegal in UK, while competely normal here. And vice-versa, in slightly smaller(?) proportion...
Anyway, better to look on the bright side of life as always as possible. :-)
So I can use wihout any fear class II medical device.

Joined: Sep 2010
Posts: 17
Novice
Offline
Novice
Joined: Sep 2010
Posts: 17
This error usually means the mains earth on the point you are using is open circuit.
We have had this happen several times and has been usefull in identifying building wireing faults.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

I have another 601-Pro here that I thought was working OK ... but when I went to CAL the test lead (across the red and green jacks) now the darned thing keeps saying something like "lead resistance is beyond limits" (or some such thing). I've tried various decent leads, but all with the same sorry result. The trouble is now I can't use the unit to test Earth Resistance of a bit of kit under test. frown

Originally Posted By: Sean Fearon
I have a lot of experience with repairing this model range, and even with full schematics you can end up chasing round in circles due to poor relay contacts (trust me on this one).

Any ideas on this one, Sean (or anyone else)? smile


If you don't inspect ... don't expect.
Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
Hello Geoff
Clean the 4mm sockets with cotton buds carefully (ensure that the tips dont break off as they are difficult to remove).
Check for approx 6.0 Volt AC across enclosure and earth terminals during Cal procedure. If this is okay your cable resistance exceeds 170 milli Ohms, and that is why it wont Cal.

Hope you get somewhere mate

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Thanks for that, Sean. I'll have a go later on. smile

Bearing in mind what you were saying about relays etc., I was also wondering if something inside could have been "shook up" a bit in transit, as the thing only arrived this morning (courtesy of Parcel Force*). But I haven't lifted the lid for a gander inside yet.

* Other (many, in fact) courier services may be available.


If you don't inspect ... don't expect.
Page 2 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 5,583 guests, and 194 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,290
Posts74,597
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5