Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 4 1 2 3 4
#46710 07/06/10 9:51 AM
Joined: Apr 2010
Posts: 52
Scholar
OP Offline
Scholar
Joined: Apr 2010
Posts: 52
Hi,
How can i convert bpm to frequency (Hz)?

Joined: Jun 2009
Posts: 796
Likes: 13
Philosopher
Offline
Philosopher
Joined: Jun 2009
Posts: 796
Likes: 13
Mmm, I don't think you can, they are not the same thing.

That being said, any complex waveform (ecg or whatever) can be expressed as a set of frequencies via a process called a "fast fourier transform" (at least my rather dim memory thinks so).
See here as a starting point
Now, the physicists and mathematicians amongst us will be better placed to advise but.......

As an aside, if you want to convert most things, try this . It's great and free !!

Last edited by DaveC in Oz; 07/06/10 10:49 AM.

Thoughts and information provided on this forum are mine and mine alone and do not necessarily reflect the policy of NSW Health. They may also be complete bollocks!!
Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

In general:-

Code:
               speed of light (m/s)
  Frequency  = --------------------  Hz
                  wavelength (m)
                       
                        1
  Period     = --------------------  s
                  frequency (Hz)

               speed of light (m/s)
  Wavelength = --------------------  m
                  frequency (Hz)
                  
  ...where the speed of light is 2.998E+8 m/s
  (for most calculations 3E+8 is close enough)
  (notice that this is 300,000 km/s)
 

You could say that a heart rate of 60 BPM is 1 Hz (that is, 1 beat per second) ... 120 BPM 2 Hz, etc. But that would be an unusual choice of terminology (in the clinical context). smile


If you don't inspect ... don't expect.
Joined: Apr 2010
Posts: 52
Scholar
OP Offline
Scholar
Joined: Apr 2010
Posts: 52
But if i have a patient simulator that generates for example a pulse wave with 30 bpm how can i verify if is reading right, there's no measure equipment that reads bpm smile my idea is to convert to frequency and verify if it is correct...
1 Hz is 60 BPM is wrong in medical context?

Joined: Jan 2005
Posts: 768
Philosopher
Offline
Philosopher
Joined: Jan 2005
Posts: 768
Easiest way is to use a stopwatch and count!


Sometimes You Can't Make It On Your Own.
Joined: Jun 2009
Posts: 796
Likes: 13
Philosopher
Offline
Philosopher
Joined: Jun 2009
Posts: 796
Likes: 13
or to print and then use an "ECG ruler". This will convert the distance between the R waves into a BPM.


Thoughts and information provided on this forum are mine and mine alone and do not necessarily reflect the policy of NSW Health. They may also be complete bollocks!!
Joined: Mar 2008
Posts: 135
Expert
Offline
Expert
Joined: Mar 2008
Posts: 135
Divide bpm by 60. Then from f = 1/t you get your answer. In your example for 30bpm:
30/60 = 0.5
1/0.5 = 2Hz

Joined: Aug 2007
Posts: 306
Master
Offline
Master
Joined: Aug 2007
Posts: 306
As has been stated by some.

1hz is a complete cycle of a wave within a second.(both positive and negative portion)

Therefore the higher the frequency (number of cycles) the greater the hertz.

If something beats at 60BPM thats one hertz. Something that beats at 120Bpm is 2Hz. 30Bpm is 0.5Hz.


Joined: Apr 2010
Posts: 52
Scholar
OP Offline
Scholar
Joined: Apr 2010
Posts: 52
for mike burns 30 Bmp is 2Hz, but bcaslisle 30 Bmp is 0,5Hz...
Who is right? smile

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Originally Posted By: Mike Burns

1/0.5 = 2Hz


1/0.5 does give 2!

BPM/60 equals beats per second. That is, a frequency of beats per second. Think about it! smile

120/60 = 2.0
60/60 = 1.0
30/60 = 0.5


If you don't inspect ... don't expect.
Page 1 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (Geoff Hannis), 3,518 guests, and 12 robots.
Key: Admin, Global Mod, Mod
Newest Members
Tekie 77, Szymon1990, Janet, james_on_ebme, Biomed Egypt
10,366 Registered Users
Forum Statistics
Forums26
Topics11,256
Posts74,517
Members10,366
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5