Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 3 of 4 1 2 3 4
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Nice one, Mate (wish I had thought of it myself)! smile


If you don't inspect ... don't expect.
Joined: Jan 2005
Posts: 45
Technologist
Offline
Technologist
Joined: Jan 2005
Posts: 45
Dear friends!
To my experience, I found, a number of instruments returned after repairs with out connecting the earth terminal inside the unit (either get disconnected or loosely connected)But while doing functional test, they may be well. But it is our prime duty to test the electrical safety test, particularly cardialogy equipments.
Inaddition to the CMIS programme, electrical safety test has to be done for the safety of patient and operator after every service.


A.Thiyagarajan,
Joined: Oct 2006
Posts: 364
Likes: 38
Sage
Offline
Sage
Joined: Oct 2006
Posts: 364
Likes: 38
I don't want to open the proverbial can of worms......but surely thats the whole point in applying the new EN62353.

Malcolm.
ps.After more than 30 years of electrical safety testing I don't think I've chalked up 50 failures(most of which resolved by replacing the mains lead)

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Let the can be re-opened! Some of us sadder ones on here love this stuff, after all!

But who are you replying to, Malcolm? Chris (at his original post) or Rajan? Different questions, I'm afraid!

But, either way, let's not wait for the implementation of this or that standard, but carry on doing the right thing as we have always done:- should EST be carried out following a repair? Yes, of course. Biomed is all about patient safety. Nothing less. smile

PS: "proper" EST includes intelligent examination of the mains cable and plug (not to mention the equipment under test itself). And (in my book, at least) checking the fuse ratings as well!


If you don't inspect ... don't expect.
Joined: Oct 2006
Posts: 364
Likes: 38
Sage
Offline
Sage
Joined: Oct 2006
Posts: 364
Likes: 38
Hi Geoff,
I agree whole heartedly,I've been
implementing the '353 since January this year and find this norm far more comprehensive than just extracting the EST from the 601.
Thank the lord our Italian plugs dont have fuses!!!!!

Malcolm.

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72



At least 62353 is readable!

But I should imagine that many NHS biomed sheds will simply carry on with their testers set on "601 auto" as usual!

Don't you just weep when you hear kit groaning under "Mains on Applied Part(s)"? frown


If you don't inspect ... don't expect.
Joined: Oct 2010
Posts: 2
Newbie
Offline
Newbie
Joined: Oct 2010
Posts: 2
Leaving aside the issue of failing IEC mains leads, which could be addressed by regular replacement, is there still a need to safety test modern medical equipment or do we continue to do this out of habit.

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Welcome to the forum, Guffy. smile

The question you pose is one that comes up on the forum quite regularly. About once a year, I would say.

Habit, perhaps. But a good habit, I would suggest - especially if carried out with a bit of thought, not to mention eyes open!

Anyway, I can't see it being kicked into touch these days, what with the 'elf 'n' safety brigade being so much to the fore! smile


If you don't inspect ... don't expect.
Joined: Oct 2010
Posts: 2
Newbie
Offline
Newbie
Joined: Oct 2010
Posts: 2
Hi Geoff

As a department we spend a lot of time performing safety tests on all our equipment at least once a year.

Is there any documented evidence to suggest that these tests need to be done yearly or could the tests be done every 2 or 3 years.

Some manufactures have changed their service intervals to either two years (Draeger) or 3 years (Covidien) unless these companies assume that we will perform the tests in between their visits.

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

There have been many earlier discussions about all this.

If you find the time, check out the Search facility.

You will come across threads like this ... and this, for example.

OK ... just one more (and a bit more light hearted, too). smile

Issues surrounding IEC cables have also been well aired.


If you don't inspect ... don't expect.
Page 3 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (Huw, vitapablo), 7,798 guests, and 448 robots.
Key: Admin, Global Mod, Mod
Newest Members
Czr, Jimmy Li, AndrewS1, Lam, Danny1621
10,388 Registered Users
Forum Statistics
Forums26
Topics11,274
Posts74,579
Members10,388
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5