Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 5 1 2 3 4 5
#49797 21/10/10 5:35 PM
Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Hello everyone
I have a problem with the LOGIQ 500 cpu card (12). When switched on, the system loads the first two notches colored below the word LOGIQ, and shows the number 2135131-2 (error number?). I want to try to repair the card. It features a Motorola MVME147-010 main board (16 MHz, 4MB RAM, 1PIO, SCSI).
I removed the battery of the timekeeper to measure it, and it is low, I think the same mother board has lost the BIOS settings. Can anyone help me? Thanks in advance

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
can anyone help me?

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
hey... my friends... Can anyone know the error 2135131-2 ? Please help...

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Sorry to hear that no-one has been able to help so far. smile

I doubt that is an error number. My guess would be that it is the unit's serial number, or some such thing. Either way, it sounds like you have either lost some basic settings, or have a corrupted hard drive or something like that.

What else is going on ... is the unit booting up?

Anyway, you may need some expert assistance. See your PM's.


If you don't inspect ... don't expect.
Joined: Feb 2009
Posts: 1,908
Likes: 18
Hero
Offline
Hero
Joined: Feb 2009
Posts: 1,908
Likes: 18
The main problem that we have had with GE Ultrasound machines, is that when we have had a software problem and the unit is down, once any of the boards has been removed and then returned we get the message that the license needs to be renewed/upgraded. I think it is GE's way of stopping un-authorised service of their machines.


I am not Flippant, I am Smart
Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Yes, you've got to safe-guard those "key codes" (strings, or whatever it is they call them).

It's the same sort of deal with most of the ultrasound units. We can only but shudder to imagine what matters will be like when such kit is "no longer supported" in a few years time. Just about the same time, of course, as they will be turning up on the pre-owned and/or donated kit scene!

But hey, who knows ... perhaps at that time the gougers will do the decent thing and release all the info into the Public Domain! Dream on. frown


If you don't inspect ... don't expect.
Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Dear Geoff and Neil,
thanks for the reply.
In fact now I'm sure I was right.
I was able to connect via remote, the machine also confirmed:

VENUS *** SYSTEM RESET / START ***
*** *** BBRAM contents are corrupted
VTSH>
VTSH> *** VENUS *** SYSTEM READY
PWRD: rt_reqprdmsg error [1100th]

BBRAM : Battery Backup Ram

Now I just have to figure out how to enter the system ID of the machine. Have you by chance P9030TB Logiq 500 proprietary manual?

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Originally Posted By: Neil Porter
The main problem that we have had with GE Ultrasound machines, is that when we have had a software problem and the unit is down, once any of the boards has been removed and then returned we get the message that the license needs to be renewed/upgraded. I think it is GE's way of stopping un-authorised service of their machines.



Hello Neil, maybe this happens on newer machines. The Logiq 500 CPU boards and I have always changed with no problem.

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Originally Posted By: Geoff Hannis

Sorry to hear that no-one has been able to help so far. smile

I doubt that is an error number. My guess would be that it is the unit's serial number, or some such thing. Either way, it sounds like you have either lost some basic settings, or have a corrupted hard drive or something like that.

What else is going on ... is the unit booting up?

Anyway, you may need some expert assistance. See your PM's.


You are right Neil, the machine has lost the data on BBRAM. This is the message I get through the terminal:

*** VENUS SYSTEM RESET/START ***
*** BBRAM contents are corrupted ***
VTSH>

Has anyone had the same problem?

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Dear Friends,
we have resolved the problem. It's very simple. GE ask 8.000 euro for the PCB. 3 hour of my work, and 6.75 euro for the battery from farnell et voilĂ , the CPU work. How strange this world... Saluti !

Page 1 of 5 1 2 3 4 5

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 3,326 guests, and 16 robots.
Key: Admin, Global Mod, Mod
Newest Members
Tekie 77, Szymon1990, Janet, james_on_ebme, Biomed Egypt
10,366 Registered Users
Forum Statistics
Forums26
Topics11,256
Posts74,519
Members10,366
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5