Originally posted by ALEJANDRO:
I have a problem with a Nihon Kohden Cardiolife defibrilator, ERROR 3.
Alejandro,Sorry for the late reply.
1. Error 3 can be cleared by re-boot the unit.
a.Turn unit power OFF by the
i) Main power switch
ii) Energy Select Dial
Wait for 5 seconds
b. Turn the unit power ON by the
i) Main power switch
ii)Energy Select Dial
If the ERROR message is still present
c. Press the CHARGE Key while pressing CAL KEY. If the Error message is still present go to Step 2.
2. Its internal relay RLY 501 in High Voltage Board may be faulty.
Error 1,2,3 & 4 will appear if your unit is kept near X ray machine, generator or high speed motor so keep unit away from high noise level surroundings.
PM me if you required more assistance.