Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
#50828 09/12/10 4:13 PM
Joined: Oct 2006
Posts: 93
Mike_d Offline OP
Adept
OP Offline
Adept
Joined: Oct 2006
Posts: 93
We have a Tosca 500 unit in a patients home in the community and we would like to be able periodically download data from it remotely any ideas.

Mike_d #50832 09/12/10 9:08 PM
Joined: Jan 2001
Posts: 42
Likes: 3
Technologist
Offline
Technologist
Joined: Jan 2001
Posts: 42
Likes: 3
Hi Mike

Are you aware of Download 2001 from Stowood Scientific. This programme is designed for downloading from TOSCA 500 and other such devices. If you want any more info please give me a call.

Dave Mulvey
Salford Royal Hospital (Hope Hospital)
Tel: +44 (0)161 206 64865

Mike_d #50833 09/12/10 9:36 PM
Joined: Oct 2006
Posts: 93
Mike_d Offline OP
Adept
OP Offline
Adept
Joined: Oct 2006
Posts: 93
Hi Dave,
Thanks for the reply we have the software the problem is that the Tosca is out in the community about 50 miles from the nearest hospital. We were thinking about been able to download to a memory card and then sending by email. I am not sure if you can download direct from the Tosca to a memory card without a laptop.

Mike_d #50842 10/12/10 11:33 AM
Joined: Oct 2006
Posts: 93
Mike_d Offline OP
Adept
OP Offline
Adept
Joined: Oct 2006
Posts: 93
I have found out that the Tosca has a RS423 port which does output an ascii code. Is there any type of data logger that could be connected to the output to store the code and then sent to us for downloading so we could analyze the code using the stowood software?

Mike_d #50843 10/12/10 11:56 AM
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

It's an interesting question, Mike. And there could be any number of ways of doing it. You could spool to a PC, for example (or, if you still have one, a Good Old BBC Micro - easy enough using one of those)!

But how would you go about triggering the data stream, I wonder? Or would it always be available? And ... how much data (how long would it take to fill up 256 MB - or 4 GB - of memory)? More information required, I'm afraid! smile


If you don't inspect ... don't expect.
Mike_d #50846 10/12/10 12:19 PM
Joined: Oct 2006
Posts: 93
Mike_d Offline OP
Adept
OP Offline
Adept
Joined: Oct 2006
Posts: 93
Hi Geoff, looks like an interesting prospect just wondering if the Tosca will be able to trigger the recording or does the data logger take care of that. Will need to do a bit of googling tonite any rs423 experts out there let me know this would be a great option.

Mike_d #50848 10/12/10 12:32 PM
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

If you look at the OEM version (and go for the P-option) it looks like you can simply switch Start/Stop by connecting/disconnecting power. That's the one I think I would go for.

The standard model would need someone to press the On and Off buttons. But you would need to be careful, as Playback mode uses those same methods (eg, holding down the On button for more than a second).

Then you would have to figure out a way of downloading that data remotely!

Meanwhile, I'm wondering about the capabilities of folk at the Tosca end, and whether they would enjoy becoming involved in data capture themselves.

It sure sounds like fun to me! smile

PS: we talked about RS-232 before!

PPS: I've just noticed a V3.5 on eBay for the "Buy It Now" price of £ 180.


If you don't inspect ... don't expect.
Mike_d #50896 13/12/10 12:15 PM
Joined: Oct 2006
Posts: 93
Mike_d Offline OP
Adept
OP Offline
Adept
Joined: Oct 2006
Posts: 93
Hi Geoff, contacted the company on that rs232 data logger but what i thought would be the easiest part has now proved to be a stumblying point converting rs232 to rs423. The rs232 datalogger will not work when you have a converter (RS232 to RS423) that requires power from the rs232 lines. I have searched the net and have not found anything that will convert and uses and external power source. Maybe some of the other contributors might have seen something or have better knowledge of the RS232, 423 configuration.

Mike_d #50898 13/12/10 12:55 PM
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Pity. You say RS-423, but I'm wondering if the kit I mentioned earlier would still capture the data. It would all depend upon the details of the serial output from the unit. smile

I believe there are doodads available to convert RS-423 (levels) to RS-232 and vice versa.


If you don't inspect ... don't expect.

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 11,200 guests, and 324 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,287
Posts74,592
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5