I'm going back a few years here - but I used to work in the computer industry where we assembled and tested circuit boards for big mainframe computers. We occasionally got boards back for repair which had been contaminated (usually with tea or coffee !) and which we cleaned and re-tested. Soap and water worked well and was perfectly safe as long as you dried the boards off thoroughly, preferably in a warm drying cabinet. We always then washed them in special cleaner which evaporated off completely - RS and Farnell sell similar products in spray cans.
There's absolutely nothing wrong with giving electronics a good wash and rinse as long as you observe some simple rules.
Disconnect ALL batteries and back-up power devices before washing.
DON'T wash boards with relays or other mechanical devices unless your absolutely sure they are fully sealed and there's no way water can get inside. Alternatively, you can remove these devices and replace them when you've finished.
Make sure everything is completely dry before re-assembling or replacing batteries etc. and use a proper PCB cleaning fluid as a final rinse.
Our Quality Control system never found any evidence that boards that had been washed had a shorter life expectancy or a worse reliability record than boards that hadn't. I suspect that the quality of the printed circuit board material and the packaging of the components is better now than it was 20 years ago, so there should be even less chance of causing problems.
We haven't had to take drastic action with many medical devices and some have been beyond saving anyway because the contaminating fluid has obviously been in the device for some time and has caused corrosion.
I agree with the comment about the MS26 - they are remarkably rugged devices and tolerate a lot of abuse !