I believe I may have cracked this one.
It turns out that the GW's are not faulty at all (well, not in this regard, at least). Indeed, if there is any fault here then it really lies with the operator (or, in this case, the tester), I’m afraid.
It seems that the giving set needs to be properly and fully primed before attempting to run (test) the pump. Both upstream and downstream conditions need to be right when the mechanism "pulls back" during start up, as this is indeed when the pump checks for a "Bad Set".
It is worth stating also (and obviously, perhaps) that the correct giving set needs to be used (and a fresh one at that). It also seems a bit fussy about the 300 mm or so head height (distance of the bag above the pump) as well. In short, the test has to be properly set up.
I have found that running the Volumetric Accuracy Verification test (Code 5) described in the manual ensures that all is well before embarking on a final test using the pump analyser (IDA
et al).
Sometimes (often, in fact) persistence pays off!