Any transformer will experience a high inrush current at switch on and the bigger the transformer the bigger the inrush. Almost all isolated supply will be protected by an MCB with a type B rating. Ideally, if large inductive loads are being used, a type C rating would be expected.

The problem is even if the transformer is fitted with in-rush protection this is usually via a NTC (high resistance when cold and low resistance when hot). So once the unit is powered on the NTC has a low resistance and is effectively bypassed. If now the equipment is quickly turned off and on then the in-rush protector resistance will be low so a high current will occur. This can happen when plugging equipment in or out, as the contacts in the plug may make and break rapidly. Remember it can take a few minutes for the NTC to cool fully and restore its effectiveness.

The same applies for most switch mode power supplies so these can also be a new source of problems!

Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Sounds like a good enough reason for a GFCI at every isolated outlet. smile

Although I'm now wondering if the real answer isn't RCBO:- "Residual-current Circuit Breaker with Overcurrent protection"*. I'm not sure if these are available for single circuits (individual outlets), however.

Originally Posted By: Mr.Google

MCB Types:-

Type B are designed to trip at fault currents of 3 to 5 times rated current (In).
For example a 10 A device will trip at 30 to 50 A.

Type C are designed to trip at 5 to 10 times In (50 to 100 A for a 10 A device).

Type D are designed to trip at 10 to 20 times In (100 to 200 A for a 10 A device).


I wonder what became of Type A?

1) Source

2) Many others are available.

* An RCBO is a combination of an MCB and RCD and provides both overcurrent protection and earth fault current protection in a a single device.

Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Things are getting (even more) complicated now Mike! smile

OK, the in-rush limiter is basically a NTC thermistor. Where does it actually sit in the overall circuit (of the type we're discussing)?

Do we now need yet another component ... an "in-rush limiter cooler"? How about using a Peltier device? think ... I'll get on it straight away! Ha, ha.

Maybe in the future we'll be looking at real-time monitoring of individual supply circuits (outlets) - via computer algorithms I guess - with all sorts of protection device options set at various tripping levels (and response times) depending upon specific applications (and, set up by program variables - or maybe just good old DIP switches)! Expensive though (and let's hope they also remember to include fuses as the line of last resort)!

By the way, "in-rush" may well have been the problem here (as well as in many other mysterious - and often "one-off" - faults on high-current (switching) equipment).

Joined: Feb 2007
Posts: 300
Likes: 16
Master
Offline
Master
Joined: Feb 2007
Posts: 300
Likes: 16
Originally Posted By: Geoff Hannis

OK, the in-rush limiter is basically a NTC thermistor. Where does it actually sit in the overall circuit (of the type we're discussing)?
Clearly any NTC in this case sits at the input (primary) of the mains isolation transformer (or often any large transformer). It is also possible to perform this function by using a resistor that is bypassed by a relay contact after a time delay. It is also possible to have more sophisticated electronic devices that reduce transformer in-rush.

Please note that it is not allowed to install any such device in the primary of a medical IT system (IPS) transformer!

As for complex monitoring, there are moves by the manufacturers to try and make such individual supply monitoring compulsory for medical locations! The question no one can answer is what is a "normal" monitored level and what is abnormal? If a socket-outlet one day has a residual current of 1mA and the next 2mA is this dangerous or normal? Will such equipment enhance safety or simply drive up costs? Already we have lots of additional IPS units that may be doing nothing but often the infrastructure behind these units is not all it should be. For example I know of cases where all the hospital power went off for hours, even the IPS supplies! The doctors carried on via torchlight!

Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Good points; and good questions! smile

Maybe it would be a wiser course to have any complex monitoring incorporated (where need be) into the mains input side of the individual items of equipment themselves.

On your last point:- sounds like (yet) another case of lack of maintenance (which should include regular testing) of the standby generator(s). frown

Page 4 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 14,077 guests, and 24 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,249
Posts74,482
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5