I have 2 units of Mindway VS-800 which both have NIBP & SpO2 function. But now, both units were found a module not working.
One of the unit is the SpO2 function not working and the other is the NIBP function not working. And I can't see the module version info. from the unit like it is not build to have....
I tried to cross check the modules (the boards) within these 2 units and found that the not working module can works well on other unit, both of them!!
Any idea of what's going wrong? Is the Main board problem? Thank you so much!
Try a Reconfig - these were factory instructions when i repaired a couple a few years ago. 1. Install the recorder into the VS-800 monitor. 2. Hold on the menu key for 1 second. 3. Then hold on the power key together with the menu key simultaneously for more than 2 seconds and release both keys. 4. Check the LCD display and you will find the prompt message below: "Clear system config and restarting". 5. The monitor will restart and show the following prompt message: "Checking system config…" 6. After a successful configuration, there will be a prompt message showing the successful configuration. "System config succeed, restarting..." Then the unit will restart and you can operate.
Thanks for your instructions but I can't follow to see "clear system config and restarting" after step 3... The LCD display just like a normal startup screen....