Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
#65927 22/09/13 10:40 AM
Joined: Feb 2007
Posts: 303
Likes: 17
MikeX Offline OP
Master
OP Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
During medical electrical safety testing is performing 500V DC insulation tests of any value in enhancing safety?

Do you use the 62353 alternative method for leakage measurements?

Please vote and in a months time we can debate the results!

Insulation Testing
multiple choice
Always perform DC insulation tests. (75%, 9 Votes)
Only perform insulation test if liquid damage is suspected. (8%, 1 Votes)
Only perform insulation tests if device has heating elements. (8%, 1 Votes)
Never perform insulation tests. (8%, 1 Votes)
Total Votes: 12
Voting on this poll ends: 22/10/13 10:20 AM
62353 Alternative testing
single choice
I always use the 62353 alternative method where possible (11%, 1 Votes)
I only use the alternative test to check safety before I use a direct or differential method. (11%, 1 Votes)
I only use the 62353 alternative method if I suspect liquid damage in place of an insulation test. (0%, 0 Votes)
I never use the 62353 alternative method. (78%, 7 Votes)
Total Votes: 9
Voting on this poll ends: 22/10/13 10:38 AM
MikeX #65928 22/09/13 10:57 AM
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Mike, how come you missed out:-

"Only perform insulation test if the kit keeps blowing fuses"?

And (by the way), where has this apparent emphasis on "liquid damage" come from? think

OK ... my votes have been cast. But I may as well come right out now and say that (in my opinion) the insulation test is a fundamental part of routine electrical safety testing. Other "fudges" can sometimes be used to reassure us that all is well (ie, when you don't have an insulation tester to hand), but nothing actually beats a proper 500 VDC test. smile

Or are you suggesting that all those (many thousands) of "Meggar tests" have been in vain? frown

Joined: Jul 2005
Posts: 601
Philosopher
Offline
Philosopher
Joined: Jul 2005
Posts: 601
Has anybody ever carried out "Evidence Based" research into EST? From my experience, other than the mains lead we never get any faults.
I've had some in the past fail the Insulation test, mainly mineral insulated heaters and a few very old Kontron Micromons where the mians inlet filter failed.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

[quote=biomedbill]

othe

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

... what I thought I had posted before the site took an extended tea-break was:-

Originally Posted By: biomedbill

other than the mains lead we never get any faults.


Yes; electro-medical equipment generally has a mains lead [sic]! whistle

Cable!

Joined: Jun 2007
Posts: 98
Adept
Offline
Adept
Joined: Jun 2007
Posts: 98
Well Biomedbill, your experience is borne out by PAT statistics. The mains lead / IEC lead is the major source of danger. (I have occasionally failed items on an acceptance test when part of a complex system, the manufacturers were not happy but modified the equipment to pass the IEC test in use at the time.) However in general very little fails which is good in one sense that most medical equipment is electrically safe when in use. The complication and danger is (in my experience) that the iec mains leads do not get checked and it is surprising how many are in use with their earth wires disconnected.


Barney
Joined: Feb 2007
Posts: 303
Likes: 17
MikeX Offline OP
Master
OP Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
Originally Posted By: Geoff Hannis
Mike, how come you missed out:-

"Only perform insulation test if the kit keeps blowing fuses"?

And (by the way), where has this apparent emphasis on "liquid damage" come from?

With hindsight I should have asked " If you perform insulation tests only when there is signs of damage or evidence of liquid ingress." Hopefully people will apply this revision when answering the question.

Originally Posted By: Geoff Hannis
Or are you suggesting that all those (many thousands) of "Meggar tests" have been in vain? frown

Well I don't want to bias the outcome of the poll so I will reserve comment until after all the votes are in!

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

You're right, Mike; perhaps we should keep our powder dry. smile

But let me just respond to Barney, if I may ...

@Barney: how can you test the kit without its "lead"?

Or do we have folk who carry a "known good" IEC-320 mains power cable around with them as part of their electrical safety testing gear? think

The "lead" is part of the test! And "mains lead management" is part of the job.

If you had someone going around testing only IEC mains power cables then that person would be gainfully employed, I reckon. And that includes checking the mains plug (even if it is "moulded") and fuse on each "lead" as well, by the way! whistle

Joined: Jun 2007
Posts: 98
Adept
Offline
Adept
Joined: Jun 2007
Posts: 98
Geoff, you ask : how can you test the kit without its lead?

You test the kit with a lead (more often an iec lead) but that lead is very unlikely to stay with the same equipment. I have witnessed persons testing class 2 equipment with the iec lead and pass both after only a class 2 test. Often the iec lead has an earth conductor and should be checked independently as a class 1 item, since obviously it will pass a class 2 test with a dissed earth and then if used with another piece of class 1 equipment render it unsafe.

My opinion is it is more important to test the iec leads and plugs say 6 monthly whilst most class 2 equipment (and some class 1) could be tested at a much longer interval.

Last edited by Barney; 26/09/13 5:06 PM.

Barney
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

I hear you Barney. smile

As you stroll around the hospital, test every IEC-320 cable you can get your hands on (even if it's just with your pocket multimeter); and always carry a pair of cutters to "remove" every one that needs to be "binned". Also bin any and every "Class II" mains cable (that is, those without an earth conductor); they are literally "accidents waiting to happen" (an immediate SFC)!

Perhaps I should forget about (early) retirement and re-emerge into and upon unsuspecting hospitals and continue to preach to, moan (cuss) at, or otherwise cajole those "young pups" (today's young biomeds) into not only doing, but also thinking about what they are doing when it comes to electrical safety of electro-medical equipment (and by extension - sometimes literally - that of the patient).

But who's going to pay me for traveling around? Or should I be doing it as a (yet another) "labour of love"? think

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 14,130 guests, and 32 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5