Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 2 of 2 1 2
#6053 12/01/07 2:46 PM
Joined: Mar 2001
Posts: 182
Mentor
Offline
Mentor
Joined: Mar 2001
Posts: 182
Hi Chaps and Chappesses ??

No we're not considering repairs as Clement Clarke will not supply us any parts. We'd certainly save a mint if we could replace all our bent arms!

But before we had the prism and rod, every time a tonometer was dropped it had to be sent back to CC for checking/calibration at £300 at time. At least now we can check to see if damage has occurred. If it's only a calibration issue then we have the specs to work to and bring it back into cal.

Interestingly, we check returning tonometers from CC after repair and they're sometimes out of cal, more than likely due to the TLC they receive at the hands of the post office. These are returned again and eventually they come back and agree with Haag's specs.

The calibration adjustment available is limited to the movement of the counterweight under the top cover. The securing screws are quite small and obviously with a precision instrument undue force is not recommended. - Contrast that to the way they are handled by the ophthalmic guys!

If you still can't achieve calibration then there may be some distortion of the mechanism or contamination of the bearings which will need to be resolved by CC.

However limited, it does allow significant savings to be made along with minimised downtime.

You might also talk to your ophthalmic staff and persuade them into buy a mounting arm for the tonometers which can be swung out of the way when not required. Since these remain permanently attached to the slit lamp there is minimal handling of the tonometer and we have certainly benefitted from them.


All the best

Brian

Hope this helps.

Brian


**** Still grumbling in the Pennines ****
#6054 15/01/07 9:36 AM
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
Hi Brian

I do not want to carry out major repairs on these Tonometers, we have the pressure arm and a prism so are able to do the instrument checks and would like to be able to carry out adjustments when required.

The problem that we frequently see is that the zero point has moved (so instead of -0.05 and +0.05 you have 0 and +0.1). Quite often this is due to the arm being bent, when it is not I presume the counter weight requires adjustment however with no information/instructions we can do not carry out these adjustments.

Lee


Don't forget "we've never had it so good".
#6055 15/01/07 9:38 AM
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
Sorry there was a spare "do" in the last line of my previous reply.

Lee


Don't forget "we've never had it so good".
#6056 15/01/07 11:23 AM
Joined: Jul 2001
Posts: 87
Adept
OP Offline
Adept
Joined: Jul 2001
Posts: 87
Good old google has produced a clinial paper, authored in the uk in 2005 reporting that "Goldmann applanation tonometers are not as accurate as the manufacturer's recommended calibration error tolerance of +/-0.5 mm Hg would suggest. Calibration error of less than +/-2.5 mm Hg is clinically acceptable. Calibration error checks should be carried out once monthly and tonometers with calibration error greater than +/-2.5 mm Hg returned to the manufacturer for re-calibration"
Do we need this accuracy?? any comments??

Bob..

#6057 15/01/07 11:45 AM
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
Bob

This takes us back to the old argument, should you follow the manufacturers instructions or not, and I believe the consensus is that if you don't you better have a good reason (in case of future litigation).


Lee


Don't forget "we've never had it so good".
#6058 16/01/07 12:25 PM
Joined: Feb 2004
Posts: 14,802
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,802
Likes: 72
Yes, Bob. The user should carry out such checks as often as possible. Especially, for instance, after dropping the tonometer on the floor! When it comes to adjusting them (ie, by us), it’s hardly a high-tech bit of kit, is it? smile


If you don't inspect ... don't expect.
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
Further to this thread from a few years ago.

Frank lists "Inami - General Maintenance for Tonometer Maintenance manual" under ophthalmology.

If you take a look towards the end of the manual you will find a surprisingly familiar looking Tonometer with instructions on adjustment for calibration.

Lee


Don't forget "we've never had it so good".
Joined: Jan 2016
Posts: 2
Amm Offline
Newbie
Offline
Newbie
Joined: Jan 2016
Posts: 2
HI TO ALL TECHNICIANS,

JUST WANT TO ASK OTHER THAN HAAG STRIET, ARE THERE ANY OTHER COMPANY WHO DOES THE REPAIR AND CALIBRATION OF THE T900?

AMM

Page 2 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 5,318 guests, and 38 robots.
Key: Admin, Global Mod, Mod
Newest Members
Janet, james_on_ebme, Biomed Egypt, Nektarios K., Keisya Andretta
10,364 Registered Users
Forum Statistics
Forums26
Topics11,254
Posts74,506
Members10,364
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5