Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 3 of 3 1 2 3
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Ariel, perhaps do you have the user or service manual for a Maquet HS22K5?

Joined: Sep 2010
Posts: 412
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 412
Likes: 2
No need to mention on me, if I ever have I'll send it immediately. smile

Have you ask our friend google?


--- PATIENT SAFETY FIRST ---
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Yes, google doesnt have it, very difficult ot find it.

Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
My unit is now working fine on the 121°C programas.
When we choose the 134°C programs, temp and pressure rises up to 124°C in a correct relationship, then it starts to fall, temp. reaches 134 but pressure is around 30psi below what it should be, so the unit gives a too long time error, as I imagine, is waiting for the pressure to be the correct one. I will try to look for pressure leaks, or bad valves may be??? Any toughts are welcome.

Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Door seal? think


If you don't inspect ... don't expect.
Joined: Sep 2010
Posts: 412
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 412
Likes: 2
Clog? bad valves?


--- PATIENT SAFETY FIRST ---
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
Dear friends: Finally solved all the issues with this unit, so first of all thank you all for your colaboration and always good will. Problems found:

1. No drying cycles were taking place, and if chosen only a dry cycle the unit will block.

Solution: 1. Damage fuse carrier
2. Defective memory on the main PCB AJUNC3

2. Lost of pressure: 1. The seal on the door was with a crack, changed it for a new one.
2. Exhaust Valved worm out, changed with a new one.

3. Unit now working OK

Best regards to all of you.

Joined: Sep 2010
Posts: 412
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 412
Likes: 2
What happened? how come the memory on the PCB got deffective? think


--- PATIENT SAFETY FIRST ---
Joined: Apr 2013
Posts: 44
jbt Offline OP
Technologist
OP Offline
Technologist
Joined: Apr 2013
Posts: 44
I dont know. We took this unit from the customer after they had a bad experience with the technical service of Tuttnaer representative in the country. Maybe some kind of electrical pike??

Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

There's a lot of them about here, too. wink


If you don't inspect ... don't expect.
Page 3 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (MatonSRS), 14,532 guests, and 12 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,249
Posts74,483
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5